When using a GCMS duel column, when using one column at 1.2ml/min what should i do for the second column if it's on a deans switch (do i keep at same flow or lower? my service engineer has said drop it to 0.1-0.3ml/min but im scared of dmg to column
Michael1993 , please set it at 2 psi (constant pressure mode) for the second column. Make sure that you configure the column correctly, inlet is either front inlet or back inlet, outlet is either MSD or other (0 psi).