tune action stopped: cannot achieve constant peaks widths

Hi evryone

Please help: i just get these 2 messages which make me i can not inject from my 7890A and 5973C MSD " tune action stopped: cannot achieve constant peaks widths  " also "limits fail: pump oils (days), tune time (hours) at the same moment the Hi VAC is 0.00e00

Please any help to make my instrument run? thanks 

Parents
  • Hi,

    Either call agilent to do PM or you can clean source and change rough pump oil...I faced this problem earlier but i always keep one ion source assy and EM Horn in my lab stock to change both immeditely after vent and keeping old ion source in stock again after cleaning..

    Basically, if EM horn value is more than 2500 then better to replace it along with cleaning or changing of ion souce...

Reply
  • Hi,

    Either call agilent to do PM or you can clean source and change rough pump oil...I faced this problem earlier but i always keep one ion source assy and EM Horn in my lab stock to change both immeditely after vent and keeping old ion source in stock again after cleaning..

    Basically, if EM horn value is more than 2500 then better to replace it along with cleaning or changing of ion souce...