Ghost peaks and sudden baseline shifts

5975C MSD and 7890 GC. I am fairly new to GC-MS and am having issues with my baseline. I have recently begun seeing ghost peaks and sudden baseline shifts after replacing a column. I did inlet maintenance and added an in-line gas purifier as well as conditioned the column to help resolve, but the bottom spectra which was run after this maintenance is actually showing the ghost peaks again as well as the sudden baseline shifts I saw before. 

Could this be a detector issue? I have a CI detector but know nothing about how to clean or troubleshoot it. 

Parents
  • Not really. I've listed the ghost peaks and the 18 and 40 abundance below. Please scroll over to the right to see all the numbers.

    I did run an air/water check this morning and my N2 values are quite high. O2 and water seem okay. Not sure where high N2 would come from.

    Forgive my ignorance, but what's the significance of m/z 40? I get that 18 is water.

    Thanks for the help.

     

    R.T. 18.440R.T. 20.431R.T. 23.574R.T. 25.681R.T. 33.717R.T. 36.266
    m/zAbunm/zAbunm/zAbunm/zAbunm/zAbunm/zAbun
    18.124627218.1337918.1213218.1263118.1371918.14275
    406741640934724075032407610440811284084760
    R.T. 38.685R.T. 42.9544R.T. 43.729
    m/zAbunm/zAbunm/zAbun
    18.1454618.1538518.15602
    408606440822564080672
Reply
  • Not really. I've listed the ghost peaks and the 18 and 40 abundance below. Please scroll over to the right to see all the numbers.

    I did run an air/water check this morning and my N2 values are quite high. O2 and water seem okay. Not sure where high N2 would come from.

    Forgive my ignorance, but what's the significance of m/z 40? I get that 18 is water.

    Thanks for the help.

     

    R.T. 18.440R.T. 20.431R.T. 23.574R.T. 25.681R.T. 33.717R.T. 36.266
    m/zAbunm/zAbunm/zAbunm/zAbunm/zAbunm/zAbun
    18.124627218.1337918.1213218.1263118.1371918.14275
    406741640934724075032407610440811284084760
    R.T. 38.685R.T. 42.9544R.T. 43.729
    m/zAbunm/zAbunm/zAbun
    18.1454618.1538518.15602
    408606440822564080672
Children
Was this helpful?