5975C GC/MS 219 ION HIGH

After cleaning the source on a 5975C GC/MS and performing atune, the 219 ion is high, roughly equal to the 69 ion, and sometimes greater. The response of 69 is normal (219 is high, 69 is not low). This instrument is used for volatiles analysis and now will not pass the required bromofluorobenzene tuning criteria, since the 173-177 bfb ions are too high relative to the bfb 95 base ion. 219 cannot be lowered enough through manual tuning. The source, quad and transfer line are at the proper temperatures. Any solution to this? Try re-cleaning?

Parents
  • quick question, is there flow going into the MS?

     

    Sometimes if the vacuum is too strong it can cause higher masses to get higher responses. So if the flow is low or non-existant then this could be the cause.

     

    Or if the method it's tuning on isn't your Standby.M method, or the correct Standby.M method. This has happened a few times to me before giving out of the ordinary tune results.

Reply
  • quick question, is there flow going into the MS?

     

    Sometimes if the vacuum is too strong it can cause higher masses to get higher responses. So if the flow is low or non-existant then this could be the cause.

     

    Or if the method it's tuning on isn't your Standby.M method, or the correct Standby.M method. This has happened a few times to me before giving out of the ordinary tune results.