Manual Firmware Installation

Former Member
Former Member

Is there a way to manually load firmware onto a smartcard (instead of Agilent Configuration)? I have encounter an issue where someone has tried to update firmware using a computer that has various firewalls enabled causing the MSD to now become unresponsive.

 

Kind Regards,

 

Aden

Parents
  • I know this was covered in the first answer but we have run into this enough times I thought I would offer an endorsement.

    On certain MS we have run into this problem often. It has been caused by the corporate network connection being given first priority. Disabling the  corporate network connection and resetting the MS allows it to finish the update process. The corporate firewall software has also caused update issues and needs to be disabled before starting the update. Our IT guys, after trying their magic, have yet to find a permanent fix for prioritizing the network connections on a permanent basis.

  • I have the same problem with 5973N GCMS, i have certainly put Bad firmware with 2 lan card configured

    So how do you have found a valid firmware ? I try another way with other version but Always the same error, MSD runing again other and other. I have IP valid ping. How Can i force the firmware ?

    Best regards