What settings should I use for our ECD when not in use. I am guessing that it would be best to remove the column and cap the detector inlet. What temp and makeup gas flow is good for short term storage?
What settings should I use for our ECD when not in use. I am guessing that it would be best to remove the column and cap the detector inlet. What temp and makeup gas flow is good for short term storage?
This instrument is used mostly for FID analyses and the second injector is not working. That's the primary reason why the column will be disconnected most of the time. Also to reduce contamination of the ECD, but it sounds like that isn't a huge concern. Thanks for the suggestions.
This instrument is used mostly for FID analyses and the second injector is not working. That's the primary reason why the column will be disconnected most of the time. Also to reduce contamination of the ECD, but it sounds like that isn't a huge concern. Thanks for the suggestions.
Hello sassman ,
I wouldn't necessarily say it isn't a concern, as you would need to keep air and moisture out of the ECD if it's not in use.
For reference, the ECD manual is always a good reference tool. Pages 16 and 18 are referring to your current question, but it's good to have for future reference too.