Do you know how to set the baseline for the chemstation software? I would like to integrate the peaks like below, but....
the autointegration actually integrate with weird baseline. Do you know how to modify this baseline?
Do you know how to set the baseline for the chemstation software? I would like to integrate the peaks like below, but....
the autointegration actually integrate with weird baseline. Do you know how to modify this baseline?
Open the GC Parameters in the Method & Run Control view. In the Signals section, I believe there is an option to Zero the signal at the start of the run.
I'd also recommend going into the TCD B settings and switching the Negative Polarity setting to the opposite of what it currently is. That will invert the signal so the peaks are positive rather than negative.