change retention time with Gas Saver Time

Dear Colleagues,

I have GC-FID 8890 and i'm using Gas saver feature but I'm observing when I change time of gas saver , the retention time is changed.

for instance, I set gas saver at 20 ml/min after 2 min then inject methanol and the RT is 4.74 but when I change time to 15 min, the RT of methanol become 7.9 min!.

I'm very confused what's the relation between gas saver and retention time ??.

The column flow 2ml/min.

Parents
  • Hi, 

    Are you running split or splitless inlet conditions?  What is your total inlet flow at the time of injection?   Have you tried to run the split vent trap diagnostic to see if you have some restriction in the line?

  • Hi Dear,
    my mode is split and its flow 200ml/min.
    total flow is 205 ml/min.
    how i can run this diagnosis ?

  • I will try to turn off gas saver and kindly allow me tell you what's the result.

  • Of course!  I look forward to hearing your results.  I think we are getting somewhere Slight smile

  • Hi Abbey,
    I did some trials as the following :
     gas saver is on ( 20 ml/min after 2 min ) and column constant flow at 4ml/min
    - the RT for methanol is 4.56 min.
    gas saver is off and column constant flow at 4 ml/min.
    - the RT for methanol is 7.610 min.
    gas saver is off and column constant flow at 5ml/min
    - the RT for methanol is 5.626 min.
     gas saver is off and column constant flow at 7ml/min
    -the RT for methanol is 3.679 min.
    Since this issue is appeared, i observed the response is doubled ( before issue the response of methanol is around 5000 but when this issue is appeared the response of methanol become around 16000 !)

  • Hi Abbey 

    I did some trails as the following 

    1- gas saver is on ( 20ml/min after 2 min ) and the column flow is constant at 4 ml/min, so the rt for methanol is 4.561 min.

    2- Gas saver is off and the column flow is 4 ml/ min, so the rt for methanol is 7.610 min.

    3- gas saver is off and the column flow is 7 ml/min to achieve 5psi, so the rt for methanol is 3.679 min.

    Since this problem is appeared, I observed the response is increased by more than doubled ( before problem the response of methanol is around 5000 but after problem the response become 16000)

  • Good morning!  Thanks for running these experiments.  I have a few suggestions and follow-up questions.  

    Were you able to run the split vent restriction test I mentioned earlier in the thread?  What were the results?

    I suggest also running the Leak and Restriction test from the same menu. 

    Have you recently changed liners?  What is the part number of your current liner?  

    Did you see the response of your methanol peak change with the increased column flow?  I would also be interested to see how the RTs behave if you run 3 repeat injections under the same conditions.   

    Do you have access to a flowmeter and/or leak detector?  

  • Abbey, I hope you're doing well and am grateful for your follow-up.

    i implemented what you recommended.

    Were you able to run the split vent restriction test I mentioned earlier in the thread?  What were the results?

    yes, sir 

    for the split vent restriction test, I opened the split valve ( that I circulated in the attached photo ) then I did the test and passed.

    I suggest also running the Leak and Restriction test from the same menu. 

    I also did it and passed.

    Have you recently changed liners?  What is the part number of your current liner?  

    No, I never changed the linear and for type and part number, kindly review the attached photo.

    Did you see the response of your methanol peak change with the increased column flow?  I would also be interested to see how the RTs behave if you run 3 repeat injections under the same conditions.   

    for the response of methanol to the increased column flow, I attached the results files, and the flow value is known by sample name and considering normal = 4ml/min.

    for measuring repeatability of injections, kindly review the attached result files.

    Do you have access to a flowmeter and/or leak detector?  

    unfortunately, I don't have.

Reply
  • Abbey, I hope you're doing well and am grateful for your follow-up.

    i implemented what you recommended.

    Were you able to run the split vent restriction test I mentioned earlier in the thread?  What were the results?

    yes, sir 

    for the split vent restriction test, I opened the split valve ( that I circulated in the attached photo ) then I did the test and passed.

    I suggest also running the Leak and Restriction test from the same menu. 

    I also did it and passed.

    Have you recently changed liners?  What is the part number of your current liner?  

    No, I never changed the linear and for type and part number, kindly review the attached photo.

    Did you see the response of your methanol peak change with the increased column flow?  I would also be interested to see how the RTs behave if you run 3 repeat injections under the same conditions.   

    for the response of methanol to the increased column flow, I attached the results files, and the flow value is known by sample name and considering normal = 4ml/min.

    for measuring repeatability of injections, kindly review the attached result files.

    Do you have access to a flowmeter and/or leak detector?  

    unfortunately, I don't have.