change retention time with Gas Saver Time

Dear Colleagues,

I have GC-FID 8890 and i'm using Gas saver feature but I'm observing when I change time of gas saver , the retention time is changed.

for instance, I set gas saver at 20 ml/min after 2 min then inject methanol and the RT is 4.74 but when I change time to 15 min, the RT of methanol become 7.9 min!.

I'm very confused what's the relation between gas saver and retention time ??.

The column flow 2ml/min.

Parents
  • Hi, 

    Are you running split or splitless inlet conditions?  What is your total inlet flow at the time of injection?   Have you tried to run the split vent trap diagnostic to see if you have some restriction in the line?

  • Hi Dear,
    my mode is split and its flow 200ml/min.
    total flow is 205 ml/min.
    how i can run this diagnosis ?

  • Hi AhmedIsmail, 

    On the front screen of your GC, click the house icon to go to the Home view.  Under the Diagnostics tab, go to the Diagnostic Tests section, and locate the inlet you are using.  Split vent restriction test should be an optional test.   See picture below from one of my systems (this is a remote capture, so yours may not look identical).  

    In your case, Gas Saver should decrease the split vent flow from 200 to about 15mL/min after the injection is complete, keeping the septum purge and column flows constant.  There isn't a good reason to continue flushing the liner with 200mL/min of gas after the sample contents are on the column, so 2 minutes is usually plenty of time - especially for something as light as methanol.  Gas Saver should not impact the retention time of your compound significantly unless there is a leak or restriction in your inlet.   Do you notice any alerts or not ready conditions while the run is in progress?   Are you using an autosampler?

  • No Sir I didn't observe any alerts or not ready whether before or after processing.

    I utilise manual injection and syringe in good status and clean.

  • Are you using Prep Run before you make your injection?  Auto-samplers enable this feature as part of a pre-injection readiness, which creates a more consistent inlet environment before the injection.  If there is no auto-sampler, you will need to enable the feature for best performance.  Your software should prompt you by saying 'Waiting for Prep Run'.  On the 8890 touch screen, the icon looks like a vial with a loop.  Its located in the lower right of the screen.  

    We want the conditions of the inlet to be the same for consistent results.  However, if you are not using Prep Run with manual injections, your Gas Saver conditions are still being applied during your next injection, and held until the Gas Saver time is reached.   Gas Saver conditions will be maintained until the GC is triggered to prepare for the next run to save gas while idle.  

    I would encourage you to collect a diagnostic signal for inlet total flow within your method (if you aren't sure how to do this, respond with the type of software you are using, and I'll do my best to assist).  

    For the conditions you have provided thus far, it should look like:

    Prep Run = split flow is closed off, so you should see a small amount of flow (maybe 5mL/min based on the conditions you have provided)--Correction:  This will happen in splitless mode.  Split mode behavior is dependent on if you had Gas Saver enabled on the previous run or not.  Sorry for the confusion! 

    Start of run = 205mL/min

    Gas Saver Start time (2min) should see a very quick drop to 20mL/min for the remainder of the run

    After the temperature program has completed and your run is complete, you should still see a flow of 20mL/min until you hit Prep Run again to prepare for your next injection.  Try this a few times and see if your retention time for methanol is more consistent.   Let me know what happens!

    Abbey

Reply
  • Are you using Prep Run before you make your injection?  Auto-samplers enable this feature as part of a pre-injection readiness, which creates a more consistent inlet environment before the injection.  If there is no auto-sampler, you will need to enable the feature for best performance.  Your software should prompt you by saying 'Waiting for Prep Run'.  On the 8890 touch screen, the icon looks like a vial with a loop.  Its located in the lower right of the screen.  

    We want the conditions of the inlet to be the same for consistent results.  However, if you are not using Prep Run with manual injections, your Gas Saver conditions are still being applied during your next injection, and held until the Gas Saver time is reached.   Gas Saver conditions will be maintained until the GC is triggered to prepare for the next run to save gas while idle.  

    I would encourage you to collect a diagnostic signal for inlet total flow within your method (if you aren't sure how to do this, respond with the type of software you are using, and I'll do my best to assist).  

    For the conditions you have provided thus far, it should look like:

    Prep Run = split flow is closed off, so you should see a small amount of flow (maybe 5mL/min based on the conditions you have provided)--Correction:  This will happen in splitless mode.  Split mode behavior is dependent on if you had Gas Saver enabled on the previous run or not.  Sorry for the confusion! 

    Start of run = 205mL/min

    Gas Saver Start time (2min) should see a very quick drop to 20mL/min for the remainder of the run

    After the temperature program has completed and your run is complete, you should still see a flow of 20mL/min until you hit Prep Run again to prepare for your next injection.  Try this a few times and see if your retention time for methanol is more consistent.   Let me know what happens!

    Abbey

Children
  • Dear Mr. Abbey,

    I'm using Pre run before infection .I was collecting diagnostic signal for my total flow inlet like you recommended and i observed after completing the condition of column, the gas saver value persist until i active the Pre run, the total flow is increased from value of gas saver 20 ml/min to total flow value 205 ml/min ( not like you state as below )

    (maybe 5mL/min based on the conditions you have provided)

    i attached some screen shots during my process.
    1- column parameter 

    2- before injection, Pre run and after conditioning the column 

    3- after using Pre-run

    4- after elapsing 2 min and active gas saver 

    Kindly, let me know your feedback.

  • Hi,

    Apologies, but I am not sure what the Gas Saver settings are.  Still 20mL/min at 2 min?  Were all of these screen shots during the course of a single run?   Did your peak elute at the proper time?   Has there consistently been a ramped flow on your column?  This option affects the prediction and interpretation I provided in earlier messages.

    In your ChemStation editor (where you were to grab these screen shots), under Signals, locate an unassigned signal line, and using the drop down, choose the inlet total flow - actual.  This will create a signal you can open in your data analysis to see what the total flow looks like throughout the run.  You could also select the column flow parameter to see how the parameters interact during the run without having to open the method editor.  

    As a side observation, your inlet pressure is very low, which can cause some interesting behaviors.  The 8890 system controls better at low pressure than its predecessors, but I like that number to be at least 5psig at the beginning of a run.  If you don't have the flexibility in the chromatography to increase the pressure, you may consider selecting a column with equivalent phase ratio in a smaller ID.  

  • Yes,sir the gas saver is still 20ml/min after 2 min and when change the time to 10 min,the RT is changed like I mentioned before.

    Yes, I use ramped flow in column at 2ml/min for 1.2 min then ramped 10ml/min per min till reache 4ml/min.

  • I was trying to use constant column flow and the issue persist.

  • Thanks for clarifying.  Of note, at the time your Gas Saver is enabled (2min), you have just finished a ramped flow increase, so your column flow at that point is 4mL/min, not 2.  The diagnostic trace will tell you when the system is stable at the new flow.  I would recommend waiting until that time before you turn on gas saver, like 2.5 or 3 minutes.  Alternatively, is there a chromatographic reason why you cannot simply start your run at 4 mL/min?   That would simplify the inlet conditions - at least for troubleshooting. 

    A few other things I want to make sure I'm understanding correctly.  Under a 2min Gas Saver, your RT is around 4.7min.  When you delay the Gas Saver, your RT is later, but still elutes before the time you enabled Gas saver (RT of 7.9 min with Gas Saver enabled at 10 or 15 minutes).  This indicates that Gas Saver isn't creating the problem, but is masking it.  If you turn Gas Saver off completely, when does your MeOH peak elute? 

  • I will try to turn off gas saver and kindly allow me tell you what's the result.

Was this helpful?