After the replacement of the EPC Wich are the test recommended for challenge the GC?
After the replacement of the EPC Wich are the test recommended for challenge the GC?
Hello Grisell ,
Can you please explain what you are trying to do? I don't quite understand what you mean by "challenge" the GC.
Also, can you please share the model of GC you are using?
Hi CM Zainea;
It is a 7890A with headspace. The EPC was replaced since the GC does not reach the pressure set point. After the EPC replacement I need to challenge the GC in order to demonstrate that the equipment is operational. Is there any test that proves that the GC is working correctly after the EPC replacement? (Ej. a pressure test or a system suitability)
Hello Grisell ,
There is a leak test you can run on the inlet. I am assuming you are using a Split/Splitless, and the procedure is below:
Using the 7890 Service Mode Leak Check (agilent.com)
Let me know if that answers your question.
Hi;
Yes you answer my question. Thanks so much.
Grisell