Will a 5061-3378 APG remote cable function the same as a G1530-60930 APG Remote cable between a 7890B GC and 7697A Headspace?
Will a 5061-3378 APG remote cable function the same as a G1530-60930 APG Remote cable between a 7890B GC and 7697A Headspace?
Hello usrbjsr3 ,
Regarding my last reply, I stand corrected, the two cables have the same functionality. Both are 9-pin connectors with work the same.
The only difference between the two cables is that the 5061-3378 is shorter (1m). The only problem you could have that the cable is not long enough. But for 7890-7697 it seems like it is long enough.
I hope this helps!
Hello usrbjsr3 ,
Regarding my last reply, I stand corrected, the two cables have the same functionality. Both are 9-pin connectors with work the same.
The only difference between the two cables is that the 5061-3378 is shorter (1m). The only problem you could have that the cable is not long enough. But for 7890-7697 it seems like it is long enough.
I hope this helps!