Regular peaks in baseline when heating column

Hi All, 

I am having an unresolved issue of ghost peaks and irregularities in my GC traces. I have tried bringing the column up above the final temperature and letting it sit with no injection. During this there are regular peaks in the bassline. These peaks fade as time goes on, but are still visible even after an hour at temperature. The bassline is also a bit noisy, fluctuating by about 0.5 pA after an hour. II replaced the septum. Haven't done any other maintenance but this column and inlet have not been used much since installed (maybe 20 runs total). Pictures of the traces are attached.

Some set up details:

Column: DB-1HT, 30m, 0.32 mm ID, 0.1 um film thickness

Cool on column inlet

Hydrogen carrier gas flowing at 1.6mL/min

Max program temp is 380. These images taken at 390.

Any help is much appreciated.

Zoomed out baseline showing temp ramp

Zoomed in once temp stabilized. See regular peaks.

Close in of bassline. Regular Peaks visible and very noisy.

Parents
  • Hello  ,

    What type of GC do you have, if I may? If it's an 8890 for example there are some Diagnostic tests built-in that you can run on the Cool On Column inlet.

    Have you recently changed gas tanks (whether carrier, FID fuel, air or makeup gas)?

    Make sure that the air gas supply is set at an appropriate pressure. Somewhere between 60-80 psi usually is good, but as a rule, gas supply pressure needs to be at least 20 psi higher than what the FID pressure needed during the run. Obviously, make sure you have enough gas left in the tank (if you have a tank, and not a generator).

    You could do the same with all the gases (inlet: carrier, detector: fuel, air and makeup gas), always good to check appropriate gas supply pressure.

    If all gas supply pressures are fine, consider adding gas traps/filters to the gas supply lines.

    inlet, column and detector maintenance are always good ideas too, but with this sort of regularity, my best guess would be either a gas supply or an electrical problem.

  • The machine is a 7890A. Haven’t changed the bottles recently but they all have plenty of gas pressure. Normally run the regulators around 60 psi. We have filters on all the gasses, but I’m not sure the last time they were changed. I’m going to try some inlet and detector maintenance today, but I’m not sure what else to do on the coc inlet other than change the septum. 

  • Hello  ,

    You could try to increase the gas supply pressure to see if it changes anything (say, 80 psi or so, but definitely not higher than 100).

    Maintenance procedures for both the inlet and detector can be found in this manual: 7890 Series GC Maintenance (agilent.com). In terms of the COC, you need to also make sure you are using the correct septum and the correct insert for the column type you are using. Page 106 shows the correct septum type based on column diameter. For the insert, see how many rings you see around it, that shows you if it's the correct insert for your column type. Page 103, table 14 shows the appropriate insert (5 rings).

  • When you are doing the COC maintenance it would also be good to cut 3-6 inches from the front end of the column.  Sometimes heavy components will stick there and cause problems later.

Reply Children
No Data
Was this helpful?