I have G1888 HS and every time when restart instrument it keep changing IP adress and I need to reenter IP adress.
is any way I save IP adress permanently in instrument.
I run this instrument with Shimadzu lab solution.
Find maintenance, best practice, and troubleshooting articles.
I have G1888 HS and every time when restart instrument it keep changing IP adress and I need to reenter IP adress.
is any way I save IP adress permanently in instrument.
I run this instrument with Shimadzu lab solution.
Dear Alpesh
There is a battery in the G1888 HS. This contains information about the HS for example the IP adress. You need to be replace this battery. Contact your local Agilent support.
All the best!
Daniel