Repeatabliity problems on HP7890B

Hi Everyone,

Please can you help me. We have been having repeatability problems on the HP7890B for a long time and can't get to the bottom of it. Analysing for VOC. I have run lots of RSD measurements (10 injections). A common thing we are seeing is excellent RSD(%) and then every so often we get an increase in area of a peak which can throw the RSD(%) our quite a bit as shown in the table below. Sometimes the result is higher and we get RSD(%) of about 7%. This one outlying result can throw a whole analysis out and there doesn't seem to be any pattern to when this one outlying result will occur. Another problem being when we look at bigger amounts in standards the result if on a bad injection can be in the magnitude of 100s of ppm difference to what it should be. the table below shows that both the Benzene and Toluene peaks increase with the same magnitude.

Method information

Autosampler Turret - G4513A

Syringe - Agilent Gold Tapered (5181

Septa 5183-4757 Bleed/Temp optimised, non stick

liner - 5183-4647 (deactivated glass wool, single tapered 4mm)

Volume 1ul

Solvent pre washes, sample washes, sample postwash all x4

FID

Temp - 250oC

Pressure 6.5PSI

Total flow 36.3ml/min

Split flow ratio 2:1

Split flow rate 22.2 ml/min

Detector

Temp - 250oC

Air Flow - 350ml/min

H2 flow - 35ml/min

Makeup N2 flow - 12ml/min

Oven

Column -DB-VRX(30mx450umx2.55u)(124-1534)

Column flow - 11.1ml/min

Pressure -  6.5PSI

Velocity - 100.6ml/min

Initial oven temp - 38oC

Constant type - Pressure

Runtime - 17.3 mins

Post run temp - 50oC

Oven programme

38oC for 9 minutes

ramp 40oC/min to 90oC hold time 0 minutes

ramp 15oC/min to 180 and hold for 1 minute

Things we have tried

Maintenance - regularly changing Syringe, liner,septa, o-ring

Checked the Autosampler turrets and run on front and back FID - same results

Run the same standards on another GC in the lab a HP6890 with same method and HP6890 RSD all ok, HP7890 still problem. Used the column from te HP6890 in the HP7890 and same problem occuring.

New Column

Had Agilent visit and check the machine and they see nothing wrong with the instrument.

Looked H2 generator, and fitted a regulator so we would get a definite steady flow of Hydrogen.

Tried Splitless and split analysis

Looked at Viscosity delays and injection techniques all look fine as no change in results.

Changed the method parameters to constant flow rather than constan pressure, still the same.

Increased method time to see if peaks elute later on. there was no peaks eluting later on.

If anyone has had similar problems or have an suggestions of how to solve this problem then I would be forever in somebodies debt as we have has this probm for around 3 years and never found the solution and because of this we have a 20 year old HP6890 running everything and a newish HP7890 sat there like an expensive brick of plastic doing nothing much.