Please I need a method for the analysis of greenhouse gases especially methane and carbondioxide. I have seen from many journals using this GC for such analysis but the method used was not specified.
Thanks
Please I need a method for the analysis of greenhouse gases especially methane and carbondioxide. I have seen from many journals using this GC for such analysis but the method used was not specified.
Thanks
Agilent has greenhouse gas analyzers available: greenhouse gas analyzer | Agilent
More details on the greenhouse gas analyzers as well as other analyzer configurations can be found in the Solutions Guide: Analyzer Solutions Guide for the Energy and Chemical Industry (agilent.com)
Hello Bigbaylow
You can check above application of greenhouse gas analysis.
Since, you looking for ch4 n co2 then you can focus on fid channel as per above applucation but methanizer need for this analysis.
Hope, it's fine for u.
Hi ISH,
Thanks for the paper, it was just what I was looking for.
If your detection limit requirements for carbon monoxide and carbon dioxide are above 100 ppm then you do not need the methanizer.