HI Community. We have problems with ignition of FID on GC7890B. After few seconds the flame goes off and the gas flow goes down. At the same time our output signal ist very high 596000. Thanks for any information.
HI Community. We have problems with ignition of FID on GC7890B. After few seconds the flame goes off and the gas flow goes down. At the same time our output signal ist very high 596000. Thanks for any information.
Hi,
Follow below link to resolve FID Ignition problem.
https://community.agilent.com/technical/gc/w/wiki/7889/troubleshooting-an-fid-not-igniting
And for output signal very high, may be spring of electrometer is not seated properly with FID Block assy.
Hi ISH, thanks for the reply. I am having the same issue as Slawa. Do you know what should the interconnect spring should look like when it is seated correctly?
I have ridiculously high output, and I suspect this is the cause.
Hi
Can you share picture of signal value and interconnect of electrometer?
Hey. Johnson we solved the Problem by removing the igniter Castor with collector nut and the spring locks Info the right place. Another Problem could be the malfunction electrometer.