We got low sensitivity in FID after performance a P.M, (we changed all consumables) and performance back out of the column . The system is a 7890A (SSL-FID-ALS)
We got low sensitivity in FID after performance a P.M, (we changed all consumables) and performance back out of the column . The system is a 7890A (SSL-FID-ALS)
Can you please share what consumables changed..Jet cleaned or changed..
Septa, liner,o-ring, gold seal, ferrules, jet, trap of split,.. Yes, the jet is brand new.
U replaced jet with same bore as it was earlier..Check collector body, insulator and gasket of FID...If possible, replace collector body of fid...
Yes, the jet is the same. I am goino to check out the FID..Thanks!
Yes, the jet is the same. I am goino to check out the FID..Thanks!