The TCD detector on our 6890 Gc shows negative peaks for water acetylene , propadiene .

We have a Gc (6890) with TCD . The Gc & the method setup has been around for years . It measures some permanent gases & some Hydrocarbons .

 

The method starts with negative polarity to detect H2 (He is the carrier ) & then the polarity is switched to detect N2, O2, Ar, CO, CO2 ,H2S, methane , ethylene , acetylene , propane , propylene , butanes , pentanes

 

Off lately the TCD is giving us issues . It shows negative peaks for acetylene , propadiene, water  sometimes H2S.

 

These peaks never used to be negative . This is a new TCD ( changed in Oct 2020). We had similar issues with previous TCD So we had to change it .I am not sure why we have started seeing this negative peaks .

We never had this issue before

 

Parents
  • Hello,

    Yes, most likely your carrier gas are contaminated. If this is the case, the TCD will truly change response. If the TCD is exposed for contaminated carrier gases for a longer time, the TCD filament can be heavily oxidized. In the old days, with TCD`s on 5890 GC`s, the TCD filament could be regenerated by heating to 400C over night (Using pure helium carrier gas!)

    BR

    Lars  

Reply
  • Hello,

    Yes, most likely your carrier gas are contaminated. If this is the case, the TCD will truly change response. If the TCD is exposed for contaminated carrier gases for a longer time, the TCD filament can be heavily oxidized. In the old days, with TCD`s on 5890 GC`s, the TCD filament could be regenerated by heating to 400C over night (Using pure helium carrier gas!)