6890 liner failure

 Resurrected a 6890 which has not been used for a while. When the chemstation is turned on, it gives message of "Limits fail: liner(injection)". Pulled out and cleaned/replaced microseal (fine), glass liner with o-ring (no physical samage) and golden seal underneath(replaced). The warning still pops up.

 

 Indeed, there seems to be significant leakage, since He is consumed quite a lot, overnight in standby mode. What else should I check? Thank you.

Parents
  • Hi edta450,

    This message is an early maintenance feedback feature that is set in the software rather than a diagnostic or physical monitoring. Take a look under the instrument menu in your software( Version and type dependent) for emf untilities and you should be able to reset the counter,

    Regards

    James

Reply
  • Hi edta450,

    This message is an early maintenance feedback feature that is set in the software rather than a diagnostic or physical monitoring. Take a look under the instrument menu in your software( Version and type dependent) for emf untilities and you should be able to reset the counter,

    Regards

    James