Please could you suggest what impact setting the column outlet pressure to ambient instead of vacuum when connected to a MSD could have?
Thanks
Find maintenance, best practice, and troubleshooting articles.
Hi,
Please could you suggest what impact setting the column outlet pressure to ambient instead of vacuum when connected to a MSD could have?
Thanks
hello camelia,
if you change the column outlet parameters, the calculation of the inlet-/column pressure changes & at the same time the column flow changes.
* if it is set wrong:
~ the separation on the column is massively affected.
~ with an MSD there is an additional problem: a too high flow (if vacuum outlet = ambient) lowers the MSD sensitivity, as it is dependent on the vacuum.
* However, there are MSD systems that have a CFT splitter, where the outlet pressure NOT is set to vacuum (and its correct)
some infos can be found:
https://www.agilent.com/cs/library/eseminars/public/Carrier%20Gases%20in%20Capillary%20GC%20Analysis_021511.pdf
other links could not be passed on
regards, M.
Thank you M for your reply and the slides. A splitter is not being used so the parameter will be corrected
Best Regards
Thank you M for your reply and the slides. A splitter is not being used so the parameter will be corrected