Where do these peaks come from?

We are running a method to analyze assay and some impurities in a pharmaceutical material. We are running under the following conditions:

Agilent HP-1 30m x 0.53mm x 3.0 um column

200 C inlet

1:1 split

5 mg/mL (high concentration to detect impurities)

2.5 mL/min flow with temp gradient with 250C bakeoff for 20min

2 uL inj

Diluent: ACN 

Liquid inject

Liner: 4mm deactivated straight through with wool (Restek cat # 20782)

Septum: Bleed/Temp Opt. non-stick 11mm (Agilent PN 5182-3413)

 

We have peaks that look like "impurities" peaks but do not increase and decrease with same magnitude as the sample concentration and are inconsistent between injections of the same sample. They are not in the ACN blank. They increase with inlet temperature increase to 250C.

I also see them (with lesser magnitude) with an injection of just hydrogen peroxide in ACN only.

These peaks are seen at 7.8  min and 8.2 min in the attached chroms.

 

Any clues where these are coming from? Interaction with injection and something in the inlet? Is there a more inert liner or septum I should be using?

attachments.zip
Parents
  • If the GC is used for more years, possibility of inlet contamination. As Gary has mentioned cleaning with Methylene chloride is good and maintaining total flow 20ml/min by rising split flow. However many laboratories, methods are developed during 5890 and 6890 GC period and following it and not ready to revise their methods. Some Non-Agilent software are not functioning for splitless mode and users need to use split mode with very low split ratio.

     

    I have observed in some old GC inlets are having lot of deposit.

    It has to be cleaned with scrapper or inlet cleaning metal wire brushes. Split vent port is also very important, it gets choked and affect the split ratio. Refer attached picture below.

    Next glass wool may be also have impact on unknown peaks or low response. Use Agilent Glass wool. I suggest even new liner remove the glass wool, sonicate it for 3 minutes in IPA to remove the powders and insert new glass wool. It will improve the performance.

    Use this brush for cleaning inlet

Reply
  • If the GC is used for more years, possibility of inlet contamination. As Gary has mentioned cleaning with Methylene chloride is good and maintaining total flow 20ml/min by rising split flow. However many laboratories, methods are developed during 5890 and 6890 GC period and following it and not ready to revise their methods. Some Non-Agilent software are not functioning for splitless mode and users need to use split mode with very low split ratio.

     

    I have observed in some old GC inlets are having lot of deposit.

    It has to be cleaned with scrapper or inlet cleaning metal wire brushes. Split vent port is also very important, it gets choked and affect the split ratio. Refer attached picture below.

    Next glass wool may be also have impact on unknown peaks or low response. Use Agilent Glass wool. I suggest even new liner remove the glass wool, sonicate it for 3 minutes in IPA to remove the powders and insert new glass wool. It will improve the performance.

    Use this brush for cleaning inlet