Hello guys
we got a new Agilent 7890B one month ago; after installing the columns it happens this random solvent broading in FRONT and BACK section.
We reinstalled the columns but the problem is still there.
What we have to check again?
Find maintenance, best practice, and troubleshooting articles.
Hello guys
we got a new Agilent 7890B one month ago; after installing the columns it happens this random solvent broading in FRONT and BACK section.
We reinstalled the columns but the problem is still there.
What we have to check again?
Hi dmarchini,
It looks like solvent hanging around in the inlet. I checked the vapor volume and its good at around 320ul . I do notice that the total flow is below what we recommend for effective use of a split/splitless inlet. Typically the inlet is happiest at 20ml/min. A couple of things to try , one at a time. Turn on your septum purge , default is at 3ml. If you really need a 10:1 split ,you could also turn on gas saver at 1 minute (or other times as you experiment) which would get the total flow to what we recommend operating the inlet at and help sweep the residual solvent away. Last would be to increase the split to get the total flow up to the 20ml, but you may sacrifice some sensitivity . I hope this helps.
Regards
James
Hi dmarchini,
It looks like solvent hanging around in the inlet. I checked the vapor volume and its good at around 320ul . I do notice that the total flow is below what we recommend for effective use of a split/splitless inlet. Typically the inlet is happiest at 20ml/min. A couple of things to try , one at a time. Turn on your septum purge , default is at 3ml. If you really need a 10:1 split ,you could also turn on gas saver at 1 minute (or other times as you experiment) which would get the total flow to what we recommend operating the inlet at and help sweep the residual solvent away. Last would be to increase the split to get the total flow up to the 20ml, but you may sacrifice some sensitivity . I hope this helps.
Regards
James
Thanx James, it was right the SEPTUM PURGE, it was unchecked and switched off; now it is enabled and I obtain nice sharp solvent peak.
Thanks!
Glad it is resolved .
Regards
James