Sinusoidal baseline.

We have a baseline that gives a sinusoidal wave that starts about 16 minutes.  We are using an Agilent 6850 GC with a split/splitless injector and an FID.  The inlet is interfaced to a 7679A HS.  The baseline is good up until 15.5 minutes then it starts the sinusoidal response.  This response is consistent from one run to the next.  We've baked the column per the manufacturer's instructions.  We've re-installed the column on the detector end and put in a new inlet liner.  Any ideas? 

 

  • Ok. Now we’re cookin’ with gas.  After changing the jet we now have an improved chromatogram:

     

     

                    BUT, we still have the sinusoidal issue.  I’ve asked the analyst to remove the collector and sonicate for a few minutes in acetone and MeOH.  HOWEVER, I was told we saw this ‘sinusoidal’ effect with this method on more than one instrument.  Can the dimethyl acetamide be affecting the stationary phase to produce this?

     

     

    Curtis [Personal information removed by moderator.]

  • Def second the blank injection idea.  See if systemic or chemical.  Almost looks like something that builds up...  like the bulk of the material comes off but the remainder builds up pressure, then releases, builds up, then releases.. etc.  Perhaps a dead volume issue in your system plumbing?

  • I agree.  It looks very much like a pressure increase then sudden release repeating over and over.  But WHY during this one particular period?  Why don’t we see it throughout the chromatogram?  This pattern is seen only over this particular time/temperature range.

     

    Curtis Barnes

    [Personal information removed by moderator]

  • Because it has to reach that temp/pressure in order for the material to appear in the first place.  Hence the reason you don’t see it prior to your large peak.  Once you’ve met that T/P “sweet spot” for this material, it is continually building up, or off-gassing, or whatever it’s doing.  If you don’t inject anything but monitor the signal as you gradually increase the oven temp manually, do you still see it?  You don’t even have to do a blank injection for this.  Just set your oven at the temp where your main peak there comes out, then gradually nudge it up.  What do you see?  If it’s something in your system, you’ll start seeing junk come off.  Though if you have a “pocket” somewhere in your plumbing, it could still be systemic but related to the chemical simply because you have to introduce it for the material to build up.

     

    If you run a series of sequential blanks, do you see the same intensity or does this diminish over time?  I had a “similar” situation where I had a contaminant I could ID but not determine the source.  My first injection had a HUGE peak, but all following runs had a significantly smaller peak, but was VERY consistent.  It ended up being residue from new gas lines that built up in my system overnight, over the weekend—while my system was dormant and cool, but released when I fired it up Monday morning.  The steady peak I saw afterward was what was able to build up in my system during the cool-down btwn runs.

     

    Have you introduced any new plumbing/gas/etc recently?

     

    [Personal information removed by moderator]

  • Hi ccbarnes5,

    I am marking this post as assumed answered due to age. If this isn't the case, please feel free to post again. If any of the suggestions were helpful or helped you resolve the issue, please mark as correct answer.

    Regards

    James

Was this helpful?