Sinusoidal baseline.

We have a baseline that gives a sinusoidal wave that starts about 16 minutes.  We are using an Agilent 6850 GC with a split/splitless injector and an FID.  The inlet is interfaced to a 7679A HS.  The baseline is good up until 15.5 minutes then it starts the sinusoidal response.  This response is consistent from one run to the next.  We've baked the column per the manufacturer's instructions.  We've re-installed the column on the detector end and put in a new inlet liner.  Any ideas? 

 

Parents Reply
  • No.  There has NOT been any change with the detector; yet.  We’re going to try changing the FID.  I’ll take a look at the collector when the jet is changed.  The issue that’s really throwing us off is that it only happens AFTER the elution of the solvent (Dimethyl acetamide), when the oven temperature gets over 160°C.  AND it’s so consistent.  Subsequent chromatograms can be overlaid and that baseline will look EXACTLY the same over that time period (about 15.5 minutes – 20.0 minutes).

     

    Curtis [Personal information removed by moderator.]