Can a column be installed (and used) crosswise, from the back injector (SSL) to the front detector (FID)?
Can a column be installed (and used) crosswise, from the back injector (SSL) to the front detector (FID)?
Hi Lorelei,
I'm not sure that it matters, but could you please tell us the model number for your GC and the version of software you're using?
Thanks. - Josh
We are using both 6890s & 7890s with OpenLab
We are using both 6890s & 7890s with OpenLab
Hi Lorelei,
Whan you perform OQ on the unit make sure that all channels are qualified. For example if you have two inlets and one detector you need to perform OQ on Front Inlet -> Front Detector and Back Inlet -> Front Detector.
Thank-you, Tomi.
Have a good weekend,
Edit - removed personal contact information