Agilent 6890 High Pressure/Low Flow - Shutoff error

I am working with an Agilent 6890, Splitless Inlet, and am having problems with the pressure maxing out and cutting off flow. The Front Inlet Low Flow Shutoff notice shows and the GC stops running. I have checked the septum/septum nut, replaced the liner, checked the gold seal, and clipped the column, none of these things fixed the problem. We cannot run the instrument for more than 2-3 minutes before it shuts off.

Parents Reply
  • Hi James,

     

    The initial pressure and flow are set at 7.8. As soon as the run is started, the pressure goes as high as the supply is set to, within a half a minute. When it reaches that point, the flow lowers to 2.3 and then zero. Then the instrument says #4 Front Inlet Flow Shutdown.

    We have not changed the gold seal recently and it has ran fine for the last 2 years, so I can't imagine the gold seal could be the issue.

     

    Sincerely,

    Brenda