Strange results from 6890N GC

Hi all

We are using GC6890N in our lab to analyze dry gas, the GC with TCD and FID and four columns   

Two weeks ago We noticed something strange in the results like a cycle.

N2 value it goes up and then goes back down.

Is there any advice to solve this problem?

 

Dheya

 

Parents
  • Hi Dheya,

    Since the CO2 area goes down (although only slightly) when N2 area goes up, I would suspect possibly a leak in the "sample-in" line. Testing with Snoop, water/IPA or a leak detector won't necessarily help because the sample in line is only pressurized while sampling. Also the gas sample is not necessarily detected by a leak detector.

    Try connecting He/H2 to the sample-in line and plug the sample-vent, then leak check all connections including the sample filter if used.

Reply
  • Hi Dheya,

    Since the CO2 area goes down (although only slightly) when N2 area goes up, I would suspect possibly a leak in the "sample-in" line. Testing with Snoop, water/IPA or a leak detector won't necessarily help because the sample in line is only pressurized while sampling. Also the gas sample is not necessarily detected by a leak detector.

    Try connecting He/H2 to the sample-in line and plug the sample-vent, then leak check all connections including the sample filter if used.