What are the factors affecting peak tailing in GC ?
What are the factors affecting peak tailing in GC ?
I didn't see a column flow listed. If the total flow on the inlet is 200 ml or above, its recommended for a split/splitless inlet to have a cross cut gold seal rather than the regular seal. The parameters supplied do not mention inlet type.
Cross cut gold seal for spli/splitless inlet 5182-9652 .
Liner information would be helpful as well.
Regards
James
I didn't see a column flow listed. If the total flow on the inlet is 200 ml or above, its recommended for a split/splitless inlet to have a cross cut gold seal rather than the regular seal. The parameters supplied do not mention inlet type.
Cross cut gold seal for spli/splitless inlet 5182-9652 .
Liner information would be helpful as well.
Regards
James