What are the factors affecting peak tailing in GC ?
What are the factors affecting peak tailing in GC ?
Hi koti.vanga,
can you attach image to this problem ,do you inject by auto sampler or LSV ?
The split ratio 200 high so check for leaks ,if not leaks in injector,column connection
try reinstall column in injector and detector side because the most peak tiling problem cause by leaks and bad instillation for column.
see below links for instillation
https://www.agilent.com/cs/library/Support/Documents/F03041.pdf
https://www.agilent.com/cs/library/Support/Documents/F03034.pdf
Dheya