i got baud rate error from 8000 diss. sampling station and syringe pump is not working. i added image of the error.
i really dont get what is wrong.
i got baud rate error from 8000 diss. sampling station and syringe pump is not working. i added image of the error.
i really dont get what is wrong.
The baud rate should be set to 9600.
On the 8000, press PRINT.
Select #5 "Set Baud Rate".
Set the baud rate to 9600.
Below is a link to the user manual describing this in more detail:
https://www.agilent.com/cs/library/usermanuals/Public/Agilent_8000_Op_Man.pdf
The baud rate should be set to 9600.
On the 8000, press PRINT.
Select #5 "Set Baud Rate".
Set the baud rate to 9600.
Below is a link to the user manual describing this in more detail:
https://www.agilent.com/cs/library/usermanuals/Public/Agilent_8000_Op_Man.pdf
Thank you for your answer, i already tried that and other settings such as programmable baud or enable syringe.
i believe it is something on board like IC Programed Eprom Assy VK8000 or mainboard itself.
If the baud rate is set to 9600 on the 8000 and it is still displaying this message, I would try another null modem cable and check all of the cable connections.