We have a 5200 with a 96 capillary array and want to convert to a 12 capillary array, the new 12 capillary array bundle will not fit into the reservoir opening and snap into place. The 96 array has a larger end vs. the 12 array bundle.
We have a 5200 with a 96 capillary array and want to convert to a 12 capillary array, the new 12 capillary array bundle will not fit into the reservoir opening and snap into place. The 96 array has a larger end vs. the 12 array bundle.
Hi Dennis,
Thank you for sharing this information. Based on the serial number listed in the screenshot you shared, you have a 5300 Fragment Analyzer. This system is only compatible with a 48 and a 96 capillary array. Therefore, you can only install a 48 or 96 capillary array on this instrument.
The extended method you are referring to appears to be DNF-471E33 and this method is compatible with our DNF-471 kit and either a 12, 48 or 96 capillary array. Here is the DNF-471 Quick Guide.
Please let me know if I can be of further assistance.