am I able to determine Ca in the protein sample with ICP-MS 7800 ( Include Helium collision and High Matrix)?

my sample is expected to contain  Calcium at the level of 20ppm.  we are trying to determine it with 7800 ICP-MS (includes Helium collision and High Matrix). 

my questions are following: 

1. can we reduce FP from 1550 W to 900W, so we can use cold plasma to test 40Ca? 

2. how do we define  "total dissolved solid" to the digested sample?

for example, sample weight 0.5g, after digested , then diluted to 50mL. is TDS 1%?

we need to evaporate digested liquid sample , then get some solid left, dilute to 50mL. Then  we calculate TDS by solid weight divided by 50. then TDS is much less than 1%. 

  • Hi Rodolfo, 

    Thank you for suggesting us to adjust sample depth and KE. 

    WE set sample depth to 8.0mm and 4V as discrimination energy . The signals increased, and S/N is around 8. the instrument setting was attached. 

      44Ca[He]
      CPS CPS RSD Average  S/N
    Blank  1654.57 5.2 1615 8.020433
    Blank  1617.89 3.5
    Blank  1646.78 5.9
    Blank  1587.88 7.9
    Blank  1570.11 1.9
    0.2ppm 12750.42 1.1 12953
    0.2ppm 12990.66 1.8
    0.2ppm 13032.91 1.6
    0.2ppm 13036.23 1.9

    There are several questions I would like to ask: 

    1. Can we use "cold plasma" by reducing RF power to determine Ca 40. , so we can obtain ppt level signals? 

    2. Is ICP-MS 7800 allowed to use cold plasma mode?  

    3. Are  there any drawbacks for the instrument or data accuracy if we use cold plasma .

    4. we can increase signals by increasing amount of sample to be digested. if we increase digested sample amount, do you think some carbon will accumulate in the instruments which will decrease the instrument sensitivity? 

    we currently digest 500 mg collagen sample with microwave. if we increase to 1 g of sample, do you think carbon in the digested solution possibly result in deposit on the detector? 

    5. as you mentioned before, if we put sample depth8.0mm, this will make instrument dirty faster. could you tell us which part could get dirty faster( skimmer cone or sample cone) , so we can prepare to clean. 

    I know I asked a lot of questions. hope you would not get annoyed.  I am a new technician for this instrument and learning staff. 

    Thank you 

    Best regards. 

    Di Xu