We have not tested nor validated the use of Oligo Pro II data on ProSize so we cannot comment on why it would work for your colleagues. As for why only a single row is showing, please collect the zip file for your data as outlined in my previous message…
At 310° C ion source temperature, the 502 will always be lower than typical. Also - please reduce the quad temperature down to 150° or maybe as high as 165° C... High quad temperature reduces the lifetime of the quad and only very few applications need…
Hi Andreja,
For calculating area % of two different peaks you must set one baseline left and one baseline right from the peaks. Then you have to enable "Information Pane" when clicking "View" in the menu bar and the results are shown as given in the…
Hi,
Yes it is right, please orientate on the last peak before the impurities start.
The exact baseline settings can be seen when enabling the peak information section below the analysis view (highlighted in green):
Manual integration limits…
We were able to configure the plate, we are good for now.
The 1290 Infinity II we have is coupled to Sciex 6500 MS and is controlled by Analyst Software using the Analyst Device Driver.
Hello,
No, you cannot use classic report styles *.frp from Chemstation in the CDS 2.x software. You can use any IR template built in Chemstation in CDS 2.x by copying the .rdl files over to the new project folders.
Marty
The software (except for MicroLab Pharma) comes with a 90-day trial period before the software needs to be registered. If you get in contact with our sales team, they should be able to get you a trial copy of the software. Find the appropriate contact…
Hello, Is it possible to get the Intelligent rinse working if the rinses occur after the analysis. Let's say I have 2min analysis time and I changed the Probe rinse of samples and stds to 120s and the Probe rinse 1-3 after the initial rinse. Then the…
This Information Applies To: Agilent OpenLab CDS (version 2.x)
Contents:
In the following procedure, you will find steps to create a multipoint ESTD calibration method with OpenLab CDS 2.X.
Steps to follow:
1. Load the Data
a. From Data...
This Information Applies To: Agilent MassHunter Quantitative Analysis Software
Issue
Chromatographic peak integration is a critical part of accurate and reproducible quantitative data analysis. The integrator type and the integration parameters...
This Information Applies To: Agilent 6200/6400/6500 Series LC/MS Systems
Issue
Instructions on how to fill the calibrant bottle on 6200/6400/6500 Series LC/MS Instruments when calibrant levels are running low.
Background
The ability to autotune...
This Information Applies To: All Agilent 6000 Series LC/MS Systems without a gate valve for ion capillary installation/removal.
Issue:
Installing the Ion Transfer Capillary after cleaning or replacement. For instructions for installing an ion capillary...
This Information Applies To: AriaMx Real-time PCR System
Issue
The saturation level is unique for each optical module after the offset of calibration factor and background calibration. The FAM, ROX, HEX, CY3 & CY5 optical modules will saturate at...
This Information Applies To: Agilent LC Systems
Issue
While it is possible to combine modules from different HPLC series such as 1100, 1200, 1260 or 1290 series to create a functioning LC system, there are some important topics to consider.
...
Fixes for the "RPC Server Unavailable" Error
Multiple Results
This Information Applies To: Agilent OpenLab CDS
Issue
When launching the instrument Acquisition screen/instrument session, there is an "RPC Server Unavailable" error.
Resolution
Accessing the Services App (Windows)
There are two ways to...
This Information Applies To: Agilent LC Pumps
Issue
If a user suspects that the gradient composition isn’t accurate and causes poor chromatography such as retention time shifts, late, early, no peaks or merging peaks. This article provides a quick...
This Information Applies To: Agilent 1290 Infinity (G4226A) and Agilent 1260 Infinity (G1367E) HiP Autosampler
Issue
Replacing the Needle Seat of the Autosampler is required when:
The needle seat is visibly damaged.
Leaks or blockages in the...
This Information Applies To: Agilent OpenLab CDS
Issue
This article discusses different types of bracketing modes and how to set up a bracketed calibration in Data Analysis for OpenLab CDS 2.x
Background
Occasionally, the instrument conditions...