The Introduction Service if purchased includes the following:
Customer guidance to reconfigure the GC from checkout to intended use, including a single injection of a QC standard for the desired method. In the absence of a pre-existing method, the…
In your acquisition sequence table for your calibration samples, is this selected?
Do you use the same name for your data analysis method over time? Are there multiple copies of it on your system? Did you start a new batch from the beginning or add…
Carefully remove the old O ring. There's a spot in the lower right hand edge that has an opening where you can very carefully pry it out. Do not scratch the aluminum manifold.
Use a solvent-dipped cotton-tipped swab to wipe out the entire O ring channel…
Here is a "best of" collection of links to manuals, tools, and other resources for Vacuum and Leak Detection Solutions .
Products are listed on the Vacuum Technologies home page. Get there by going to products on Agilent.com, choosing Vacuum Technologies…
Hello Kris, this valve is normally open and if you put a flowmeter on this vent line you should measure what is shown on the inlet Split Flow setting (in split mode) or Purge Fow (in Splitless mode). If Gas Saver is enabled and active it should measure…
The Agilent Cary 3500 Flexible UV-Vis system is a versatile instrument used for analyzing a privacy screen's light transmission properties at various angles and light conditions. It boasts a spacious sample compartment suitable for different accessories…
This Information Applies To: Agilent GC Systems
Issue
Excessive baseline noise can be introduced at different points including sample preparation, sample introduction, separation, and the detector. The excessive background can cause reduced signal...
This Information Applies To: Agilent GC Systems
Issue
Baseline instability that is not reproducible across multiple runs can be introduced at different points including sample introduction, separation, and the detector. The baseline instability...
This Information Applies To: Agilent GC Systems
Issue
Peak fronting is present in a chromatography peak when it has an excessive asymmetry with a leading edge. A normal peak is almost symmetrical.
Other chromatographic problems are identified...
This Information Applies To: Agilent GC Systems
Issue
A shift in retention time can occur for single, multiple or all peaks.
Other chromatographic problems are identified in Basic Troubleshooting for GC Systems . Regular maintenance of your instrument...
This Information Applies To: Agilent GC Systems
Issue
A change in peak size can occur for single, multiple or all peaks.
Other chromatographic problems are identified in Basic Troubleshooting for GC Systems . Regular maintenance of your instrument...
This Information Applies To: Agilent GC Systems
Issue
Peak splitting is when a single compound is shown in the chromatogram as two peaks usually unresolved, or one of the peaks is a shoulder.
Other chromatographic problems are identified in Basic...
This Information Applies To: Agilent GC Systems
Issue
This test isolates the point of contamination to the sample introduction system, which includes the gas supply, gas supply lines, and the sample inlet. It does not include the syringe or autosampler...
This Information Applies To: Agilent GC systems
Issue
Peak tailing is present in a chromatography peak when it has an excessive asymmetry with a trailing edge. A normal peak is almost symmetrical.
Other chromatographic problems are identified...
T his Information Applies To: Agilent 8890/8860/7890/7820/6850/6890 GC Systems with TCD (Thermal Conductivity Detector)
Issue
Problems observed with thermal conductivity detector (TCD) results can have different underlying causes.
Resolution...
This Information Applies To: All GCs, 6890, 6850, 7890, 7820, 8890/8860, Intuvo 9000
Issue:
This article is a troubleshooting guide for when the Flame Ionization Detector (FID) does not ignite or the flame keeps going out.
Cause:
If the FID...