The tune evaluation criteria are benchmarks that GCMS instruments typically pass.
"Designed for" and "works with" are two different things. The tune evaluation was created before the Extractor Source and High Efficiency Source. These ion sources get…
Hello, we were not able to tune the instrument at all. There was so much noise at low mass that it fails the initial quad coarse mass calibration. We think it is PEG contamination, possibly also PPG, but a number of masses correspond with sodiated PEG…
I have a G4226A autosampler with a recurring error message. Movement of needle into seat failed [EE 4714, 3204]
This error started after the autosampler had been repaired by a field engineer. The main board had to be replaced; the computer was unable…
Hey Ryan, try the following. Normally, the valve command failed is coming as a result of the position not found. Overtightened screws are a typical result if you have done maintenance before (happened to me the first time...). Tighten the screws each…
Hello,
We have an agilent 6545. Today when we are doing a mass calibration check, the masses of the negative tune ions seems to be shifted (For example 1633.9988 is now showing 1639.2109 and so on for the other ions) and mass calibration check seems…
I think that, with this version of the software, there are configuration files stored in the "instrument number folder", e.g. C:\Chem32\n I was going to suggest that you try to delete/rename the config.reg file, having closed all instances of Chemstation…
Hi, I am using Agilent 7820A GC-FID with EZChrom Elite v3.3.2. software. After the completion of each sequence, I would like to automatically generate a .text file along .dat file. So far, after the sequence has ended, I process the sequence and then…
I am trying to quantify ethyl lactate in ethanol 20% (remainder water). When I run multiple injections, the RSD is 6-8% for the ethyl lactate peak. The ethyl lactate standard should be 0.2% v/v. We typically test samples that are 95-100% ethanol and use…
Good Evening, I realize this is a bizzare question but basically what we have is a Dean Switch GCMS system and we use cut windows to determine when the best time is to switch from column 1 (inlet to switch to 3rd column - vent) to column 2 (inlet column…
Actually, you have to be sure that it's detector problem or not.
And if still problem persist and after changing lamps intensity test, wavelength test failed again then you have to call to your local agilent support to check electronics and optics etc…
This Information Applies To: Agilent G4208A Instant Pilot
Issue
This article describes how to switch the illumination of the autosampler on or off (if available) using the G4208A Instant Pilot.
Items Required
G4208A Instant Pilot for Agilent...
This Information Applies To: Agilent 7683B Automatic Liquid Sampler
Issue:
O nly use this alignment procedure if the injector hardware configuration has changed or if the Align Mode light comes on.
Agilent does not recommend performing this procedure...
This Information Applies To: Agilent MassHunter data acquisition software and MassHunter Quantitative Data Analysis Software
Issue
This article explains the difference between quantifier and qualifier ions and provides some general guidance on quantifier...
This Information Applies To: 5200/5300/5400 Fragment Analyzer, Fragment Analyzer controller software
Issue:
Over time the dynamic coating polymer on the capillaries needs to be refreshed. Failure to do so can lead to a loss of performance. Indicators...
This Information Applies To: Agilent ICP-MS (7700/7800/7850/7900/8800/8900)
Issue
Elevated doubly charged values that are greater than 3% for x-lens configuration or greater than 7% for s-lens configuration, can be caused by contamination, or an...
This Information Applies To: Agilent 6200 Series TOF and 6500 Series Q-TOF Systems
Issue
Instructions on how to correctly dilute the tune mix for 6200 Series TOF and 6500 Series Q-TOF systems depending on source type.
Background
If the Tune...
This Information Applies To: Agilent MassHunter Qualitative Analysis Software version 10.0
Issue
The Agilent MassHunter Qualitative software can generate qualitative analysis reports, which may include sample chromatogram information and sample...
This Information Applies To: Agilent MassHunter Qualitative Analysis Software version 10.0 or higher
Issue
Identifying compounds using a spectral library is essential in qualitative analysis. This article will describe how to load a data file and...
This Information Applies To: Agilent ICP-MS Systems
Issue
The Helium cell gas line on the ICP-MS may be contaminated with air and moisture in case of a tank replacement, or if the instrument was not used for an extended amount of time. Purging the...
This Information Applies To: Agilent MassHunter Quantitative Analysis Software
Issue
The Calibration Curve window tools help with setting an appropriate calibration curve for each compound in the calibration table.
The Calibration Curve window...