Hello EhsanKhedive ,
I can confirm that this feature does not appear to be implemented in Qual B.07.
The library RT is extracted in Qual B.08 and later, so if you require that functionality you would need to upgrade to the latest version of Qual.
Hello MatteoUnisa ,
Let me add to this by saying in limited testing it appears that Qual 12.0 is able to search compounds imported from CEF files. Still, I would recommend working with your local support team to confirm this for your workflow and investigate…
Hello MattSmithChemist and CBM ,
Product support has released new information on the 12.1 tune report and confirmed that the low mass ion was changed to m/z 42 for all TQ. It will only show on classic instruments if the option to 'Tune masses < 100…
Hello Dang ,
Manually integrated peaks are flagged in the batch table, and there are also indicators in the Compound Chromatogram.
You can also change the fill color of manually integrated peaks if you want that to stand out more.
There is…
Hello twitsok ,
Since the relative concentration for your QC is showing as different from the standards it is possible that the method has become corrupted or concentration values in the method were changed but the batch was not fully analyzed. If you…
This Information Applies To: Agilent GC systems with Gas Clean Filter
Issue
If you have determined that the Gas Clean Filter is saturated, it mus be replaced (see the article When to Replace the Gas Clean Carrier Gas Filter ).
The steps described...
This Information Applies To: Agilent GC Systems 8890/8860/Intuvo 9000 /7890/7820/6850/6890
Issue
General guidance on GC column flow rates, outlining the correct column flow rate for optimal compound separation depends on the column internal diameter...
This Information Applies To: Agilent GC Method Translator
Issue
Detailed instructions on translating methods using the GC Method Translator.
Background
For ideal use of cases and applications with the GC Method Translator, see the article...
This Information Applies To: Agilent Solvent Vent Calculator
Issue
The Solvent Vent Calculator estimates the Agilent solvent vapor exit (SVE) accessory valve timing based on solvent type, oven temperature, vent flow, and pressure.
Background...
This Information Applies To: Use of the Agilent GC Method Translator
Issue
How to correctly use the GC Method Translator to convert a method to suit a different GC configuration or to optimize a method.
Background
The GC Method Translator...
This Information Applies To: GC Pressure Flow Calculator
Issue
Instructions for the correct use of the Pressure Flow Calculator.
Background
The Pressure Flow Calculator uses the column dimensions, oven temperature, and detector outlet pressure...
This Information Applies To: The Agilent GC Vapor Volume Calculator.
Issue
The Vapor Volume Calculator determines the vapor volume based on the injection volume, solvent type, inlet head pressure, and the inlet temperature. This calculated vapor...
This Information Applies To: Agilent GCs
Issue
Agilent provides free GC calculators and tools to assist with method development.
Background
Tips : The calculators are available for downloading at Agilent.com .
Vapor Volume...
This Information Applies To: Agilent GC Systems
Issue
The Agilent Gas Clean Carrier Gas Filter delivers faster stabilization times for increased GC productivity, removes oxygen, moisture, and hydrocarbons from the carrier gas for MS applications...
This Information Applies To: Agilent GC systems
Issue
When a GC heater cannot reach setpoint, or taking too long to reach setpoint, the GC will enter a "Thermal Shutdown" (or similar heater error) state to protect itself. This state will turn off...