This Information Applies To: Agilent AriaMx real-time PCR system, AriaMx real-time PCR optical modules
Issue
The AriaMx has the capacity to accommodate six optical modules, and the scanning optics configuration ensures distinction between fluorescent dyes and individual samples. Agilent has not only validated the optical modules for the dyes that they are named after, but has also tested the compatibility of some other fluorescent dyes.
Resolution
Which fluorescent dyes are compatible with each of the optical modules is detailed in Table 1.
Agilent Part Number | Optical Module | Compatible dyes |
G8830-67001 | SYBR/FAM | FAM, SYBR Green, Eva Green |
G8830-67002 | ROX | ROX, Texas Red, CalRed |
G8830-67003 | HEX | HEX, JOE, VIC |
G8830-67004 | CY3 | Cy3, TAMRA |
G8830-67005 | CY5 | Cy5, Quasar 670 |
G8830-67006 | Atto425 | Atto425 |
Table 1. Optical module fluorescent dye compatibility
|
Caution: TAMRA and TET have been tested and found to work well with the existing CY3 module. Cross-talk for TAMRA to ROX is about 2%, but can be compensated with the cross-talk correction settings in the software. For instructions on how to change the cross-talk settings see chapter 2 of the Aria Real-Time PCR Software User Manual.
|
If you wish to use your AriaMx to detect a fluorescent dye that is not listed in Table 1, and therefore has not been validated by Agilent this may still be possible. However, Agilent it can't guarantee that the AriaMx can detect fluorescent dyes that are not validated. The optical module excitation and emission wavelengths are detailed in the table in the photo roll on the AriaMx product page, as well as in Table 2. To determine whether the fluorescence dye of interest might work, you can use a spectral overlay tool.
Agilent Part Number |
Optical Module |
LED Color |
Excitation |
Emission |
||
Wavelength (nm) |
Full Width Half Max (nm) |
Wavelength (nm) |
Full Width Half Max (nm) |
|||
G8830-67001 | SYBR/FAM | Blue | 462.5 ± 2 | 55 ± 5 | 516 ± 2 | 9 ± 2 |
G8830-67002 | ROX | PC Amber | 585 ± 2 | 12 ± 2 | 610 ± 2 | 13 ± 2 |
G8830-67003 | HEX | Green | 535 ± 2 | 10 ± 2 | 555 ± 2 | 11 ± 2 |
G8830-67004 | CY3 | Green | 542 ± 2 | 10 ± 2 | 568.5 ± 2 | 13 ± 2 |
G8830-67005 | CY5 | Red | 635 ± 2 | 14 ± 2 | 665 ± 2 | 13 ± 2 |
G8830-67006 | Atto425 | Royal Blue | 435 ± 2 | 40 ± 3 | 475 ± 2 | 20 ± 2 |
Table 2. Optical module excitation and emission wavelengths
Learn how to effectively operate your Agilent AriaMx Real-Time PCR System: |
For Research Use Only. Not for use in diagnostic procedures.
PR7001-1193