This information applies to: Agilent Triple Quadrupole LC/MS System, MassHunter 12.0 or later.
Issue
Set up an MS2 scan acquisition method.
Steps to Follow
From the Scan type drop-down list in the Time Segments window, select Scan ( Figure 1 ).
Figure 1 . Scan Type drop-down list
...
This information applies to: Agilent Triple Quadrupole LC/MS System, MassHunter 12.0 or later.
Issue
Set up an MS2 SIM acquisition method.
Steps to Follow
From the Scan type drop-down list in the Time Segments , select SIM ( Figure 1 ).
Figure 1 . Scan Type drop-down list
Fill the...
This information applies to: Agilent Triple Quadrupole LC/MS System, MassHunter 12.0 or later.
Issue
Set up a precursor ion acquisition method.
Steps to Follow
From the Scan type drop-down list in the Time Segments window, select Precursor Ion ( Figure 1 ).
Figure 1 . Scan Type drop-down...
This Information Applies to: Agilent MassHunter Quantitative Analysis software
Issue
The report graphics settings can be customized for parameters including range displayed and label.
Steps to Follow
Fixed graphics range settings
From the main menu, click Report > Generate to open the...
This Information Applies To: Agilent GC/MS systems using MassHunter software
Issue
Manual injections are needed for a GC/MS without an automated sampler. They are also used for troubleshooting specific issues such as no peaks, lower peaks than expected, ghost peaks, and other chromatographic problems...