Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Bioreagents
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • Gas Chromatography
    • GC/MS
    • LC/MS
    • Liquid Chromatography
    • Molecular Spectroscopy
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
Liquid Chromatography Portal
  • Agilent Knowledge Portal
Liquid Chromatography Portal
LC Articles How to change the needle seat on a G7167x Multisampler
  • LC Articles
  • LC Videos
  • LC Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -Liquid Chromatography
    • +LC Operation
    • -LC Maintenance
      • +LC Pumps Maintenance
      • -LC Injectors Maintenance
        • Exchanging the Needle Assembly on Agilent G1367A-D Autosamplers
        • How to Change the Needle on a G7167x Multisampler
        • How to Change the Needle and Needle Seat in an Agilent G7129x Vialsampler
        • How to change the needle seat on a G7167x Multisampler
        • How to Autoreference the G7167A/B Infinity II Multisampler
        • How to replace the peristatic pump on 1290 Infinity (G4226x) and High Performance Autosamplers (G1367x)
      • +LC Detectors Maintenance
    • +LC Troubleshooting
    • +LC Best Practices
    • +LC Firmware

How to change the needle seat on a G7167x Multisampler

Created by Damian O'Neill Damian O'Neill over 2 years ago | Last modified by Paubudge Paubudge over 1 year ago

This Information Applies To: All Agilent G7167x Multisampler modules


Issue:

Required when:

  • The needle seat is visibly damaged, blocked, or is leaking.
  • Leaks or blockages in the Vialsampler are observed, especially between the needle and needle seat.
  • The limit for the needle interaction EMF counter is exceeded.
  • Replace the needle as part of  the yearly maintenance procedures.

Steps to follow

Tools Required:

  • 8710–0510 - Open-end wrench 1/4 - 5/16 inch

  • Flathead screwdriver

 Parts Required:

  • G4267-87012 - High Pressure Needle Seat, 0.12 mm (PEEK)

    OR

    G4267-87020 - High Pressure Seat Assembly 0.075 mm (PEEK)

  See also: Needle & Needle Seats for HPLC

 Warning 
Warning: This procedure risks exposure to hazardous, toxic, or flammable solvents and reagents. Wear appropriate PPE as described in the material handling and safety data sheet supplied by the chemical vendor, and always follow good laboratory practice.
 Warning 
Warning: Risk of injury by uncovered needle. Do not open the safety cover of the needle station during normal operation. Wear safety goggles and safety gloves when removing the needle assembly
 Caution 
Caution: It is recommended to always exchange the needle assembly and the needle seat at the same time to prevent premature leakage. See: How to change the needle on a G7167x Multisampler

1. Finish any pending or running acquisition job.

2. Stop the flow at the pump and remove the solvent lines from the eluent bottles to avoid spilling solvent.

3. Close the shutoff valves at the pump if available on your pump module.

4. Close any online sessions you have connected to the LC system. Open Agilent Lab Advisor Software, connect to your LC system then select Service and Diagnostics>Maintenance Positions>Change Needle, Loop, and Seat.

5. Select Change Seat. If changing the needle assembly, then also select Change Needle. Click Start to start the maintenance task. Confirm the part number for the needle and needle seat listed in the pop-up configuration window match the new needle and needle seat being installed in the Multisampler (Figure 1).

  

Figure 1. Running the change needle maintenance task in Lab Advisor.

 6. Open the Multisampler front door.

 7. Disconnect the seat capillary from the injection valve (Figure 2).

  

Figure 2. Disconnect the seat capillary from the Multisampler injection valve.

 8. Slightly pull (1) the front clip that holds the needle seat assembly in position. Then carefully lift (2) the complete assembly from the holder (Figure 3).

 Figure 3. Removing the needle seat assembly from the Multisampler.

 9. Insert the new needle seat (1). Press it firmly into position (2) (Figure 4).

 

Figure 4: Installing the new needle seat assembly into the Multisampler.

 10. Reconnect the seat capillary to the injection valve (Figure 5).

 

Figure 5: Reconnecting the seat capillary from the Multisampler injection valve.

 11. It is also recommended to change the needle assembly to ensure a good needle/needle seat seal. See: How to change the needle on a G7167x Multisampler.

 12. Close the Multisampler front door.

 13. In Agilent Lab Advisor software close the Change needle/loop>End, click End, and wait until the needle assembly is in the needle park station. If using the local controller (Instant Pilot) close Change needle/seat.

 If possible, run a system pressure test with the outlet of the Multisampler injection valve blanked using the appropriate blanking nut. This checks that the Vialsampler flow path is pressure tight to a pressure higher than your application requires.

 Cause: Over time deposits can block the needle and the needle seat causing pressure issues and leaks. Also, wear and tear of the needle against the seat during normal operation eventually damages both assemblies, requiring replacement.

 

 Tip 

Learn more on how to effectively troubleshoot your 1260 Agilent HPLC with ChemStation:

Agilent 1260 Infinity LC Systems with OpenLAB CDS ChemStation Edition Learning Path available from Agilent Education

  • Maintenance
  • Multisampler
  • 1260 Infinity II
  • Analytical LC
  • 1290 Infinity II
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help