Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Bioreagents
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC/MS
    • LC
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
LC Portal
  • Agilent Knowledge Portal
LC Portal
Articles Troubleshooting a failed ASTM Drift and Noise Test for the Agilent RID
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • LC Portal Home
  • LC Portal Articles
  • LC Portal Videos
  • More
  • Cancel
  • Created :  10 Nov 2022
  • Modified :  6 Feb 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  Analytical LC Systems
  • Component :  LC Detectors
  • Product Name :  1100 Series LC System 1200 Series LC System 1260 Infinity II LC System 1260 Infinity LC System 1260 Infinity II Prime LC System 1290 Infinity II LC System 1290 Infinity LC System
  • Task :  Troubleshooting
Related Articles
  • Mobile Phase Preparation for Isocratic LC Systems

    This Information Applies To: Agilent Isocratic LC Systems, Agilent Refractive Index Detectors (RID) Issue: Mobile phase preparation method impacts retention time reproducibility between injections on the same instrument, by other operator, or similar instruments. It impacts systems that use premixed mobile phases including isocratic pumps or the refractive index detector (RID). Resolution: Usually, a contraction or expansion of the volume occurs when mixing different solvents, so the...
  • General Recommendations for Using an Agilent RID

    This Information Applies To: Agilent Refractive Index Detectors, G1362A and G7162A/B. Issue: The RID is sensitive to its environment and can experience stability problems when exposed to contamination, vibrations, mechanical shocks, or changes in temperature or pressure. Resolution: Considerations on the general set-up: To avoid cross-talk between different solvent channels, preferably use an isocratic pump when operating a RID. If it is not possible to use an isocratic pump, consider...
  • Optical Balance Procedure for the Refractive Index Detector

    This Information Applies To: Agilent G1362A and G7162A/B Refractive Index Detectors Issue: The amount of light falling on each of the light receiving diodes is out of balance. You may get the system message "unbalanced diodes". An incorrect signal, which cannot be eliminated by purging the sample and reference cell with the same solvent nor by equilibrating the system, may be caused by unbalanced diodes. The optical balance procedure is performed to correct such a permanent misalignment of...
  • Troubleshooting FID Noise and High Background

    This Information Applies To: All Agilent GC systems with flame ionization detector (FID) Issue: The FID uses a hydrogen flame to burn organic samples, resulting in increased ionization that is measured as a current from 5 to 10^6 pA. Normal FID background levels will be in the 5 to 20 pA range, with no sample present and the GC oven at low temperature. Agilent recommends a temperature ≥ 300 °C to prevent condensation damage. The detector temperature should be approximately 20 °C greater than...
  • Agilent Thermal Conductivity Detector Troubleshooting

    T his Information Applies To: Agilent 8890/8860/7890/7820/6850/6890 GC Systems with TCD (Thermal Conductivity Detector) Issue Problems observed with thermal conductivity detector (TCD) results can have different underlying causes. Resolution Check that the TCD settings are within the recommended range, then assess the instrument for the following errors/problems: Hardware Fault Error No signal and/or the GC front panel screen displays Front/Back Detector Error. Cause: Filament is...

Troubleshooting a failed ASTM Drift and Noise Test for the Agilent RID

Answer

This Information Applies To: Agilent Refractive Index Detectors, G1362A and G7162A/B


Issue:

The ASTM Drift and Noise Test for Agilent RID modules available within Agilent Lab Advisor Software fails. This article will help you identify potential causes of the test failure and how to resolve them.


Resolution: 

If you encounter a general RID baseline drift or a high noise that causes your ASTM Drift and Noise Test to fail, this may be due to:

  • A thermally unstable system or lab (HPLC below air conditioning, exposed to sunlight, heater radiation, air currents, etc.), please also check this article on general RID recommendations.
  • Solvent properties that change with time (baseline drifts may occur if the amount of water in organic solvents changes, hence affecting the reference cell).
    • In this case, flush out the contamination, using only stabilized solvents.
    • Clean the parts in flow path and allow the system to be flushed out and equilibrated: After flushing and cleaning the instrument, purge sample and reference cell from the same water bottle with pure HPLC-grade water overnight. You may use the recycling function. Repeat the test in the morning.
  • Ensure that the diodes are balanced.
  • Verify that the degasser is running and works properly.
  • There can also be issues with the system pressure or the temperature control (column oven, RID heater) that can affect the RID signal.
  • Ensure that the rest of the system works properly (pump leak tight, system pressure tight).

An acceptable noise should look comparable to the following graph (Figure 1): 


Figure 1. RI Signal during the ASTM Drift and Noise Test

The Noise is calculated according to the ASTM method as shown in Figure 2:

 
Figure 2: RI Signal Noise calculation
1. Noise = maximum peak - minimum peak, 2. Wander,  3. Time.

The time range for each cycle, dt, is typically set to 0.5 minutes in the test. The noise is calculated as the difference between minimum and maximum in these time intervals. The wander value, although not appearing in the test results, is determined as the peak-to-peak noise of the mid-data values in the ASTM noise cycles. The drift is calculated as the slope of a linear regression over the whole measurement.

 Tip 
Learn how to effectively troubleshoot your Agilent LC System:
Agilent 1290 Infinity LC Systems with OpenLAB CDS ChemStation Edition e-learning path on Agilent University
 
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel
Contact Agilent Support

Still Need Assistance?

Contact Agilent Support >

Privacy Statement
Terms of Use
Contact Us
Site Help