Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Bioreagents
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC/MS
    • LC
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
ICP-OES Portal
  • Agilent Knowledge Portal
ICP-OES Portal
Articles Cleaning the Agilent ICP-OES Snout
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • ICP-OES Portal Home
  • ICP-OES Portal Articles
  • ICP-OES Portal Videos
  • More
  • Cancel
  • Created :  10 Nov 2022
  • Modified :  8 Feb 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  ICP OES Systems
  • Product Name :  5100 ICP-OES 5110 ICP-OES 5800 ICP-OES 5900 ICP-OES
  • Task :  Maintenance
Related Articles
  • Cleaning the Agilent ICP-OES Cone

    This Information Applies To: Agilent 5000 Series ICP-OES (5100/5110/5800/5900) Issue The ICP-OES cone is the interface between the plasma and the optics via the orifice at the tip of the cone. Argon passes through the orifice of the cone to produce an oxygen free light path to the optics. Damage to the cone allows oxygen to enter the light path and reduces sensitivity at low wavelengths. Damage occurs during routine operation by build-up of material and corrosion on both surfaces of the cone...
  • ICP-OES Detector and Instrument Calibration

    This Information Applies To: Agilent 5000 Series ICP-OES (5100, 5110, 5800, 5900) Issue How to calibrate the ICP-OES ICP-OES calibration is a two-step procedure: The detector calibration performs a dark current calibration to correct for detector and electronic background signals. The instrument calibration uses a specified calibration solution to calibrate the wavelengths of the polychromator. ICP-OES wavelength calibration solution is available as either Agilent ICP-OES wavelength...
  • Routine Maintenance 700 Series ICP-OES Systems

    This Information Applies To: Agilent 700 Series ICP-OES Systems Issue The 700 Series ICP-OES requires regular maintenance, and routine maintenance tasks are able to be performed by the instrument operator. Any maintenance procedures not specifically mentioned in this article should only be performed by Agilent-trained, Agilent-qualified, or Agilent-authorized Field Service Engineers. Warning : Eye Hazard Danger to eyes. The plasma is an intense light source...
  • Calibration Failure Troubleshooting on ICP-OES Systems

    This Information Applies To: Agilent 5000 series ICP-OES systems Issue Instrument calibration (wavelength), detector calibration (dark current, or method calibration (worksheet) failure. Resolution There are several reasons why a calibration may fail. The following is a list of common areas to check if problems are encountered with calibrations. The steps to follow will depend upon whether the calibration tests start, whether all the lines in your calibration have failed, or whether...
  • Maintaining the Axial Pre-Optic Window

    This Information Applies To: Agilent 5100, 5110, 5800, 5900 ICP-OES Issue The axial pre-optic window needs checking from time to time to ensure it is clean. If the axial pre-optic window is dirty, it must be cleaned or replaced to ensure adequate sensitivity in the UV region of the emission spectrum. If the axial wavelength calibration fails for wavelengths less than 190 nm, this may indicate that the axial pre-optic window is dirty. Steps to follow Follow the steps in this article...

Cleaning the Agilent ICP-OES Snout

Answer

This Information Applies To: Agilent ICP-OES 5000 Series (5100/5110/5800/5900)


Issue

Argon passes through the snout to produce an oxygen free light path to the optics. The snout sits next to the plasma and during routine operation contaminants can deposit on the snout.


The ICP-OES sample introduction components are detailed in Figure 1.

Figure 1. Torch compartment

1. Axial pre-optics window assembly 2. Cone 3. Cone thumb screw 4. Snout 5. Radial pre-optics window assembly 6. RF coil 7. Torch

Warning
Warning: The torch compartment operates at high temperatures. Allow the compartment to cool before removing components. The snout may be contaminated with chemicals. Wear Personal Protective Equipment when handling ICP components. 

Steps to follow:

Caution 
Caution: Do not touch the RF coil when removing or installing the torch or snout. Always pull the torch straight down until the torch tube has cleared the loader before pulling it away from the instrument.
  1. Remove the torch.

  2. Remove the snout: Gently pull the snout off the radial window assembly and then slide it toward the right.

  3. Wash the snout:

    1. Wash with dilute detergent to remove any solid buildup from inside the snout.

    2. Rinse both the inside and outside with de-ionized water.

    3. Remove any remaining contaminants by wiping with a lint free wipe and rinse.

    4. Thoroughly dry the inside and outside of the snout with compressed nitrogen or laboratory grade air.

  4. Install the snout and torch: 

    1. Slide the snout in and gently push to fit onto the radial window assembly. Align so the tabs on the snout are at the bottom and horizontal.

    2. Insert the torch. Holding the torch by its base, slide up the torch into the loader as far as it will go. Slight resistance should be felt as the base of the torch engages with the loader. The slot on the torch quartz tube should be parallel with the slot on the snout.

    3. Adjust the snout: Gently slide the snout toward the torch until the end of the snout touches the wall of the torch.

Tip  Routine maintenance: Agilent recommends cleaning the snout on a weekly to monthly basis. 

Tip 
Learn how to effectively maintain your Agilent ICP-OES:
ICPOES-5110-1270e - Agilent 5110 ICP-OES Maintenance and Troubleshooting e-learning course available from Agilent education

Was this helpful?
  • Yes
  • No
Thank you for your feedback!