This Information Applies To: Agilent 7000A/B/C/D Triple Quadrupole GC/MS systems
Issue
When a triple quadrupole GC/MS system tune fails, the user can check for any abnormalities by setting the tune file to default tune file before checking for hardware contamination or other issues.
Steps to...
This Information Applies To: 5977A and 5977B
Content: This procedure is to guide the installation of a capillary column on 5977A/B GC/MS with a standard column nut or self-tightening nut.
Warning : a. The GC and MS operate at high temperatures. Do not touch any parts of GC and MS until...
This Information Applies To: Agilent 5977A/B GC/MSD with EI source, EI Extractor Source, or CI Source.
Issue:
During normal operation of the GC/MSD, the source will require cleaning. It is time to clean the source when:
There is a loss in analyte response not improving by GC inlet, or column...
This Information Applies To: Agilent HES source, Agilent 5977 Series GC/MSD
Issue:
During normal operation of the GC/MSD, the source will require cleaning. It is time to clean the source when:
There is a loss in analyte response not improving by GC inlet, or column servicing
Poor calibrant...
This Information Applies To: Agilent 5977 and 5975 Series GC/MSD systems
Issue
The autotune fails and reports the error Tune action stopped: Gain and offset = 0 and peaks are too narrow!
Resolution
The MSD autotune program is run after routine maintenance and at customer set intervals when...