Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Bioreagents
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • Gas Chromatography
    • GC/MS
    • LC/MS
    • Liquid Chromatography
    • Molecular Spectroscopy
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
Gas Chromatography Portal
  • Agilent Knowledge Portal
Gas Chromatography Portal
GC Articles How to Clean an FID
  • GC Articles
  • GC Videos
  • GC Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -Gas Chromatography
    • +GC Operation
    • -GC Maintenance
      • -GC Systems
        • Replacing the Inlet Septum for Split/Spitless and Multimode Inlets
        • How to replace the Gold Seal on the Split/Splitless (S/SL) Inlet for an Agilent GC
        • Replacing the Split Vent Trap Cartridge Filter on a 8890/8860 GC System
        • Replacing the Split Vent Trap Cartridge Filter on a 7890/7820 GC System
        • How to Clean an FID
        • Bake Out the Agilent Flame Ionization Detector
        • Bake Out the Agilent Thermal Conductivity Detector
        • Bake Out the Agilent Micro Electron Capture Detector
        • Agilent GC Firmware Update Procedure
        • How To Find More Information About the Agilent Gas Chromatograph micro-ECD Wipe Test Kit
        • How To Use and Configure Early Maintenance Feedback (EMF) on the Agilent 8890/8860 GC Systems
        • How To Clear the Early Maintenance Feedback (EMF) on the Agilent 8890/8860 Gas Chromatograph
        • Column Configuration for the 8860/8890 GC System
        • Column Configuration for the 7890 GC System
      • +GC Samplers
    • +GC Troubleshooting
    • +GC Best Practices

How to Clean an FID

Created by Anna Gärtner Anna Gärtner over 2 years ago | Last modified by Carlos Vargas Carlos Vargas 5 months ago

This Information Applies To: Agilent 6850 A/B, 6890 A/N/Plus, 7820A, 7890A/B, 8860, 8890 Gas Chromatographs with Flame Ionization Detector (FID)


Issue:

The cleaning of the Flame Ionization Detector (FID), may be required in the following scenarios:

  • Preventive or periodic maintenance.
  • Resolving performance issues of the detector:
    • High background, noise, or spiking of baseline.
    • Problems to ignite the flame. 

Steps to follow:

Tools Required: 

  • Torx T20 screwdriver (p/n 5182–3465) 
  • 1/4 x 5/16-inch wrench (p/n 8710–0510) 
  • 1/4-inch nut driver (p/n 8710–1561) 
  • Tweezers 

Consumables (as needed):

  • Depending on their condition, replace the detector consumables, e.g., FID jet, PTFE insulators, ignitor, silicone ring, etc.
  • Cleaning kits are available for more convenience:
  • FID Maintenance Kit (p/n G1531-67000)  
    Includes:
    • PTFE Collector Insulators
    • Silicone Rubber Base Gasket
    • Ignitor
    • Cleaning brushes for Collector
    • Cleaning Wires for Jet
  • FID Cleaning Kit (p/n 9301–0985) 
    Includes:
    • Reamers
    • Wire mini brushes
    • Handle

 Warning 
Warning: Cool the detector and oven to a temperature below 80 °C. Wear safety goggles when working with compressed gases.

 Caution 
Caution: Always wear clean gloves while handling any parts that go inside the GC or the detector. All inner parts of the collector assembly should be treated as ultra inert.

  1. Cool the detector and oven to a temperature below 80 °C. 

  2. Turn off the detector and uninstall the column from detector.

  3. Disconnect the ignitor. 
    Figure 1 Ignitor connection

  4. Uninstall the ignitor by turning counterclockwise with a 5/16-inch wrench. Check state of copper washer and replace if necessary. 

    Figure 2 Ignitor uninstallation
    1. Copper washer

  5. Using a Torx T20 driver, remove the manifold assembly by unscrewing the three screws in an alternating manner.

  6. Lift the collector assembly straight up. Wiggle softly while doing so to avoid damaging the interconnect spring. 

    Figure 3 Removal of collector assembly
    1. Interconnect spring

  7. Unscrew the knurled brass nut by turning counterclockwise.

  8. Remove the knurled brass nut, the castle with its washer, the upper insulator, the collector, the lower insulator, and the silicone rubber gasket at the bottom of the collector assembly. 

    Figure 4 Parts of the collector assembly

    1. Knurled brass nut, 2. Castle, 3. Slot for interconnect, 4. Spring washer, 5. Collector, 6. Upper PTFE insulator, 7. Lower PTFE insulator, 8. Silicon rubber gasket

     Tip 
    Tips: Note that the longer end of the collector faces down.

  9. Inspect, clean, or replace the parts.

    1. The collector body can be sonicated in a beaker with soft detergent in water for 5 minutes, cleaned with a brush or cotton swab. 

       

      Figure 5 Cleaning of the collector

    2. Do not sonicate the FID castle, just rinse or wipe. Otherwise the PTFE coating will be damaged. 
    3. Rinse all parts with distilled water and finally rinse with reagent grade methanol or hexane. Let all parts air dry. 

       Caution 
      Caution: Do not expose the silicon rubber ring to organic solvents as they might dissolve the material.

       Warning 
      Warning: Methanol, hexane, and their vapors are flammable. Methanol and hexane can influence health permanently. Check safety data sheets before use and wear appropriate protective clothes. Discard solvents in appropriate containers.

  10. Set up the collector assembly after cleaning: Re-insert the lower insulator, the collector, upper insulator, castle with washer and knurled nut. Tighten by hand. Slip the silicone rubber gasket at the bottom of the collector assembly. (Figure 4) 

     Tip 

    Tips: The longer end of the collector faces down.

     Tip 
    Tips: The lower and upper insulators are the same why they can be placed indistinctly. The insulators have one completely flat side and one with a border around the central hole, the border points towards the ends of the collector.

  11. To remove the jet, use the 1/4-inch nut driver and lift it with tweezers.


    Figure 6 Jet removal
     1. Jet 2. Interconnect spring
     Caution 
    Caution: Avoid touching the interconnect spring to not damage it.
  12. Examine the detector base for any particles - blow out the base with compressed air or nitrogen. (Use safety goggles.)

     

    Figure 7 Top view of detector base.
    1. Inner detector entrance.
  13. With a pipette bulb, blow from the bottom to remove all particles or debris. (Use safety goggles.) 

    Figure 8 View on the detector base. 
    1. Outer detector entrance.
  14. With an unfolded paper clip, you can remove pieces of graphite that may be stuck in the bottom of the detector. 

    Figure 9 Cleaning of detector bottom.
  15. Clean the jet with a thin steel wire (0.010 inch, can also be the string 1 of a guitar) or replace it. 

    Figure 10 Cleaning of jet with a wire.
     Caution 
    Caution: Scratches on the FID jet affect its performance. If you choose to clean the jet, be sure not to scratch or bend the jet.
  16. Reinstall the jet. (Step 11, Figure 6) If the jet is new, hand-tighten until it stops and give 1/6 of a turn with the nut driver. If it is a re-used, hand-tighten until it stops and give it a slight squeeze, then 1/16 of a turn with the nut driver. Do not tighten the jet too much as you can break it.

  17. Reinstall the collector assembly. (Figure 4 in reverse)

    1. Place it on the body of the detector and press it carefully down. (Step 6, Figure 3 in reverse) Take care not do damage the interconnector spring. The interconnect spring should “pop” into place and lay in the groove of the collector body!

    2. Fit the three screws of the manifold assembly and tighten, install the ignitor, and connect it. (Steps 3–5 in reverse)

  18. Install your column according to the Reference guide.

  19. Turn on temperatures, pressures, and flows, let the detector warm up at least to 250 °C degrees, light the flame, and test performance of your FID.

 

 Success 
The baseline signal of the FID should be around 2 to 20 pA.

 

 Tip 
Learn more on how to effectively troubleshoot your FID:
FID Maintenance and Troubleshooting e-learning course available from Agilent education.
  • Detectors
  • GC Systems
  • Maintenance
  • FID
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help