Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Bioreagents
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC/MS
    • LC
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
GC Portal
  • Agilent Knowledge Portal
GC Portal
Articles How to Bake-out the Split/Splitless Inlet (S/SL) on Agilent GC Systems
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • GC Portal Home
  • GC Portal Articles
  • GC Portal Videos
  • More
  • Cancel
  • Created :  10 Nov 2022
  • Modified :  1 Aug 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  GC Systems
  • Component :  GC Inlets
  • Product Name :  6850 GC System 6890 GC System 7820 GC System 7890 GC System 8860 GC System 8890 GC System Intuvo 9000 GC System
  • Task :  Maintenance Troubleshooting
Related Articles
  • Troubleshooting "Vacuum Limit Not Reached" Error in Agilent LC Pumps

    This Information Applies To: Agilent G1311B/C, G7110B, G7111A/B, G7112B, and G7120A Pumps and Agilent G7122A, G4225A, G1379A/B, and G1322A Degassers Issue The error message "Vacuum Limit Not Reached" will appear if the degasser does not reach a preset limit (default value is 100 mbar) in a predefined time window (default 8 minutes after powering on the module). This may happen for many reasons. One reason may be due to the condensation of vapors from volatile organic solvents. The vapors...
  • Troubleshooting a Leak or Poor Chromatography on the Agilent 7697A Headspace Sampler

    This Information Applies To: Agilent 7697A headspace sampler. Issue: The headspace sampler may present a flow shutdown or poor chromatography with decreased area counts or shifted retention times, caused by leaks or restrictions in the sampler. The hardware has a built-in restriction and pressure decay test that can be utilized to help with troubleshooting. Things to know about the test: Runs two tests successively, the restriction and pressure decay test and the crossport leak test...
  • Peak Splitting in GC

    This Information Applies To: Agilent GC Systems Issue Peak splitting is when a single compound is shown in the chromatogram as two peaks usually unresolved, or one of the peaks is a shoulder. Other chromatographic problems are identified in Basic Troubleshooting for GC Systems . Regular maintenance of your instrument in accordance with the recommended maintenance schedule will reduce the incidence of chromatography problems. Background Peak splitting and broadening are a result of the...
  • Wavy Baseline on the Agilent Bioanalyzer

    This Information Applies To: 2100 Bioanalyzer, 2100 Expert software Issue The Agilent 2100 Bioanalyzer system electropherograms in the Agilent 2100 Expert software have a wavy, uneven baseline ( Figure 1 ). Figure 1. Wavy, uneven baseline on Bioanalyzer electropherograms The error is most often associated with: Problems during chip priming Dirty electrodes Moisture on the electrodes Sample contamination The sample concentration being higher than the assay's range Liquid spillage...
  • Contamination of LC/MS Tune Mix - Diagnosis and Fixes

    This Information Applies To: Agilent LC/MS Systems Issue Contamination of the tune mix or the Calibrant Delivery System (CDS) of your LC/MS system can cause Autotune failures and drops in sensitivity due to ion suppression. Items Required Alconox (p/n 5190-1401 ), or Citranox (p/n 5188-5359 ) Ultrapure LC/MS Grade Water Ultrapure LC/MS Grade Acetonitrile Background Contamination of the tuning or reference mix used on your LC/MS system can occur in several different ways...

How to Bake-out the Split/Splitless Inlet (S/SL) on Agilent GC Systems

Answer

This Information Applies To: Agilent GC Systems with Split/Splitless Inlet (S/SL)


Issue

The Split/Splitless Inlet (S/SL) can accumulate contaminants when high boiling point and/or high molecular weight compounds are present in the injected sample. Often these compounds are present from the sample matrix. The accumulated  contamination is not effectively removed using the analysis method parameters. These accumulated contaminants may appear in the chromatogram as ghost or artifact peaks. They can also be retained on the inside surface of the inlet liner, gold seal, or inlet weldment increasingly affecting the performance of the inlet over time.

 


Background

During the S/SL inlet bake-out procedure (Steps 1 to 7), high boiling point and/or high molecular weight compounds are subjected to higher inlet temperature and flow in an attempt to vaporize and remove these contaminants. The contaminant molecules, if vaporized, are purged out of the inlet under high split flow to the split vent. A bake-out of column and detector are in the subsequent steps (Steps 8 to 12) of the procedure.

Steps to Follow

 Warning 
Warning: The oven, inlet, and detector may be hot enough to cause burns. Wear heat-resistant gloves to protect your hands.
  1. Create and save a new method for the bake-out procedure. 
  2. Set the inlet into split mode. 
  3. Set the column flow to the normal operating value, or set the capillary column gas velocity to ~30 cm/s.
  4. Set the inlet split vent flow to 200 mL/min by adjusting the split ratio. For example, if the column flow is 2 ml/min, set the split ratio to 100:1. 
  5. Set the inlet temperature to 300 °C, or 25 °C above the normal operating temperature (do not exceed 350 °C as it may thermally damage the inlet liner O-ring).
  6. Make sure to send to the instrument the settings from steps 2 to 5 before continuing.
  7. Purge the column with carrier flow for at least 15 minutes before heating the oven.
  8. Open Method Editor in the software and proceed to the next step. 
  9. If the column is attached to the detector, set the detector 25 °C above normal operating temperature (do not exceed 350 °C). If the column is not attached to the detector, cap the detector fitting.

     Caution 
    Caution: Heating the oven/column up before adequate purging time with carrier gas could damage the column since residual oxygen in the carrier gas, when heated will damage the stationary phase of the column. 
    Do not set the inlet, oven/column, or detector temperature above the column being used manufacturers maximum isothermal operating temperature.
  10. Set the column oven 25 °C above the GC method final oven temperature to bake contaminants from the column. Do not exceed the column manufacturer's maximum temperature limit.
  11. Bake out for 30 minutes or until the detector baseline is stable, and free of contamination peaks.
  12. Return the instrument to operating condition by downloading the analytical method from the software to the instrument. 

If this S/SL bake-out procedure does not restore the inlet performance, replacement of the inlet liner and gold seal will likely be required as the accumulated contamination has not been totally removed. This would be typical of samples with matrix compounds that are nonvolatile but soluble in the injected sample solvent.

 Tip 
Learn how to effectively troubleshoot your GC Split/Splitless Inlet:
GC-7890-2245s - Agilent 7890 Series GC: Split/Splitless Inlet Troubleshooting e-learning course available from Agilent education

 

 
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel
Contact Agilent Support

Still Need Assistance?

Contact Agilent Support >

Privacy Statement
Terms of Use
Contact Us
Site Help