Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Bioreagents
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC/MS
    • LC
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
Automated Electrophoresis Portal
  • Agilent Knowledge Portal
Automated Electrophoresis Portal
Articles Reservoir Vent Valve Cleaning for Parallel Capillary Electrophoresis Instruments
  • Articles
  • Files
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • Automated Electrophoresis Portal Home
  • Automated Electrophoresis Portal Articles
  • Automated Electrophoresis Portal Files
  • More
  • Cancel
  • Created :  10 Nov 2022
  • Modified :  7 Feb 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  Automated Electrophoresis Systems
  • Component :  Femto Pulse Systems Fragment Analyzer Systems ZAG DNA Analyzer Systems
  • Product Name :  5200 Fragment Analyzer System 5300 Fragment Analyzer System 5400 Fragment Analyzer System Femto Pulse System ZAG DNA Analyzer system
  • Task :  Troubleshooting
Related Articles
  • Re-ordering the Syringe for Cleaning the Reservoir Vent Valve of the Agilent Fragment Analyzer, Femto Pulse, or ZAG

    This Information Applies To: Fragment Analyzer, Femto Pulse, ZAG Issue: A new syringe is required to clean the reservoir vent valve of the Agilent Fragment Analyzer, Femto Pulse, or ZAG (see Figure 1 ). Figure 1 . Syringe (marked in red square) mounted to the reservoir vent valve of a Fragment Analyzer. For all instruments, perform reservoir vent valve cleaning when needed, for example when experiencing migration issues, or dried gel is visible in the valve. Exchange the syringe whenever...
  • Failed to Build Pressure Error on a Parallel Capillary Electrophoresis Instrument

    This Information Applies To: 5200/5300/5400 Fragment Analyzer, Femto Pulse, ZAG DNA Analyzer and Oligo Pro II instruments, and their respective controller software. Issue Failed to build pressure error comes up shortly after starting a separation. Resolution Failure to build pressure error occurs when the system is trying to build pressure when liquid is being pumped during conditioning or flushing of the instrument. Possible causes and suggestions to address the issue are: 1. Empty reagent...
  • Agilent Capillary Array Installation on Parallel Capillary Electrophoresis Instruments

    This Information Applies To: 5200/5300/5400 Fragment Analyzer, Femto Pulse, ZAG, Oligo Pro II Issue: This article explains the steps to take after installing a brand new array in the Agilent 5200/5300/5400 Fragment Analyzer, Femto Pulse, ZAG, and Oligo Pro II systems. Steps to follow Capillary array replacement steps are detailed in chapter 8 of the Fragment Analyzer user manual , Femto Pulse user manual , ZAG DNA analyzer user manual , and chapter 9 of the Oligo Pro II user manual...
  • No Data From One or More Capillary on Parallel Capillary Electrophoresis Instruments

    This Information Applies To: 5200, 5300 and 5400 Fragment Analyzer, Femto Pulse and ZAG DNA analyzer Issue When no data is available in the Agilent ProSize data analysis software for one or more capillaries, the capillary(ies) could be misaligned, blocked, require extra conditioning, or the electrodes might be encrusted. Steps to Follow Check the capillary alignment Check the alignment of the capillaries in ProSize. Checking the capillary alignment is briefly described in this section...
  • Insert Disk Error in the Agilent Fragment Analyzer Controller Software

    This Information Applies To: 5200/5300/5400 Fragment Analyzer controller software, Femto Pulse controller software, ZAG DNA Analyzer controller software, Oligo Pro II controller software Issue: The Agilent Fragment Analyzer, Femto Pulse, ZAG, or Oligo Pro II controller software does not open when the desktop icon is double-clicked. Instead, one of the following errors come up: "Insert the 'Fragment Analyzer' disk and click OK" "The feature you are trying to use is on a network resource...

Reservoir Vent Valve Cleaning for Parallel Capillary Electrophoresis Instruments

Answer

This Information Applies To: Fragment Analyzer, Femto Pulse, ZAG, and their respective controller software


Issue:

If there is dried gel in the reservoir vent valve of your Agilent Fragment Analyzer, Femto Pulse, or ZAG, cleaning is required. A clogged reservoir vent valve can impact the separation performance, for example leading to delayed migration. 

 Tip Recommended maintenance schedule:  Agilent recommends inspecting the vent valve for clogs as part of troubleshooting procedures and your monthly maintenance. Clean if necessary. For more information on maintenance, please see the preventative maintenance schedule in the respective user manual.

 


Steps to follow

  1. From the Utilities menu in the controller software, select Clean Reservoir Vent Valve. The actual reservoir vent valve opens for cleaning and also the Clean Reservoir Vent Valve dialog box will be shown on screen (see Figure 1). 
     
    Figure 1. How to open the Clean Reservoir Vent Valve dialog box
  2. Use the 10 mL Luer lock syringe to:
    1. either follow the onscreen instructions and clean the reservoir vent valve sequentially with 7 mL of DI water and 10 mL of air,
    2. or for a more thorough cleaning to remove severe clogs, clean it five times with 7 mL of 0.5 N NaOH, followed by 7 mL of DI water and then 10 mL air.
  3. For the cleaning steps, attach the filled 10 mL syringe to the Luer lock fitting of the reservoir vent valve as shown in Figure 2 and press down the plunger.

 

Figure 2. Emptied syringe indicated with red box attached to the reservoir vent valve of a Fragment Analyzer

In case you need to re-order the 10 mL syringe, please use any 10 mL Luer Lock tip plastic syringe (not sold via Agilent Technologies). The Luer lock tip is compatible to the reservoir vent valve.

 

Tip
Learn how to effectively maintain your Agilent parallel capillary electrophoresis instruments:
Fragment Analyzer system manual
Femto Pulse system manual
ZAG system manual

For Research Use Only. Not for use in diagnostic procedures.

PR7000-8101

 
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel
Contact Agilent Support

Still Need Assistance?

Contact Agilent Support >

Privacy Statement
Terms of Use
Contact Us
Site Help