Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC
    • LC/MS
    • Life Sciences Reagents
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Real-Time PCR (qPCR)
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
Automated Electrophoresis Portal
  • Agilent Knowledge Portal
Automated Electrophoresis Portal
Articles Peak Fronting in NGS Library Smears on Agilent TapeStation
  • Articles
  • Files
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • Automated Electrophoresis Portal Home
  • Automated Electrophoresis Portal Articles
  • Automated Electrophoresis Portal Files
  • More
  • Cancel
  • Created :  10 Nov 2022
  • Modified :  7 Feb 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  Automated Electrophoresis Systems
  • Component :  ScreenTapes and Reagents TapeStation Systems
  • Product Name :  2200 TapeStation System 4150 TapeStation System 4200 TapeStation System
  • Task :  Troubleshooting
Related Articles
  • How to Determine the DV200 of FFPE RNA Samples on the Agilent TapeStation

    This Information Applies To: 4200, 4150 and 2200 TapeStation, TapeStation analysis software A02.02 or higher. Issue: When using the Agilent 4200, 4150 and 2200 TapeStation systems, the DV 200 of FFPE RNA samples can be calculated within the TapeStation analysis software. The DV 200 score is a quality score for evaluating quality of RNA derived from formalin-fixed paraffin-embedded (FFPE) samples established by Illumina Inc. in 2016. It represents the percentage of RNA fragments larger than...
  • Discrepancy Between Agilent RIN and RINᵉ Values

    This Information Applies To: 2100 Bioanalyzer, 4150/4200/2200 TapeStation Issue: Discrepancy between Agilent Bioanalyzer RNA integrity number (RIN) and Agilent TapeStation RNA integrity number equivalent (RIN e ) values Resolution: First, please read the following application notes about RIN and RIN e : Performance of the RNA ScreenTape Assays on 2200 TapeStation System Performance of the RNA ScreenTape Assays on 4150 TapeStation System As can be seen in the application notes...
  • Intensity Loss for 200 nt Ladder Peak in the Agilent TapeStation RNA Assays

    This Information Applies To: 2200/4150/4200 TapeStation systems. RNA ScreenTape and High Sensitivity RNA ScreenTape assays,TapeStation analysis software Issue: The 200 nt peak of the Agilent RNA ScreenTape and High Sensitivity RNA ScreenTape Ladder has lost its normal intensity. An example of this phenomenon is shown in Figure 1 with the actual 200 nt peak highlighted with a red circle: Figure 1 . RNA Ladder with rudimentary 200 nt peak in red circle If the peak is below a certain...
  • Agilent TapeStation Software Displays a Different ScreenTape Expiration Date

    This Information Applies To: 4200, 4150 and 2200 TapeStation Systems Issue The expiration date displayed in the Agilent TapeStation controller software is different to the one printed on the ScreenTape device. Resolution The expiration date printed on the ScreenTape device applies until the device is removed from its packaging. Once a ScreenTape device is removed from its protective plastic sleeve, it dries out quicker. As the ScreenTape device dries out, bubbles can form in the gel,...
  • How to Export Agilent TapeStation Logs

    This Information Applies To: 4200, 4150 and 2200 TapeStation instruments, TapeStation analysis software, TapeStation controller software Issue: Technical support require the instrument log files to troubleshoot hardware issues with the Agilent 4200, 4150 and 2200 TapeStation systems. It is possible to export the logs via the TapeStation controller or analysis software. Steps to follow Tip : Log files do not include data files, experimental, or sample information. Collect and...

Peak Fronting in NGS Library Smears on Agilent TapeStation

Answer

This Information Applies To: 2200/4200/4250 TapeStation systems, TapeStation D1000 ScreenTape assay, TapeStation High Sensitivity D1000 ScreenTape assay, TapeStation analysis software


Issue:

A sharp peak appears at the lower end of the NGS library smear in Agilent TapeStation D1000 ScreenTape and High Sensitivity D1000 ScreenTape assays (see Figure 1).

 

 

Figure 1. Region analysis (blue borders) of an NGS library smear. The red arrow indicates the peak fronting.


Resolution: 

Peak fronting is an electrophoretic artifact. The cause is a concentration-dependent stacking effect on D1000 and High Sensitivity D1000 ScreenTape assays. In certain circumstances, some fragments might show a similar mobility and stack together, forming a distinct peak within the smear. It is more obvious in low-input libraries with medium to low concentrations. It does not represent a contamination in the library. Potential causes include factors such as the sample resuspension buffer, sample smear size, and concentration range. Therefore, rerunning the samples on the same ScreenTape kit will not help as the effect is inherent of the sample.

Peak fronting does not affect the overall quantification as long as the smear is correctly integrated (see region included in Figure 1). However, the sizing might be slightly shifted and therefore the molarity might be marginally altered as well. Nevertheless, quantification and sizing results are usually meeting the kit accuracy and precision specifications. Peak fronting typically does not alter results significantly, as long as the samples are meeting kit sizing and concentration specifications (see also Quick Guides and for smear specifications technical overview):

Kit Sizing range Quantification range (DNA fragments) Quantification range (DNA smears)
D1000 35–1000 bp 0.1–50 ng/µl 5–100 ng/µl
High Sensitivity D1000 35–1000 bp 10–1000 pg/µl 0.5–15 ng/µl

Please use the following search terms to find more information about this and other NGS library phenomena in the Agilent Information Center (AIC) on your PC: TapeStation Analysis in the SureSelect and HaloPlex Workflow.

Root cause: 
Concentration-dependent stacking effect

 Tip 
Learn how to effectively operate your Agilent D1000 and High Sensitivity D1000 ScreenTape assays:
Agilent D1000 ScreenTape Assay for TapeStation Systems Quick Guide
Agilent High Sensitivity D1000 ScreenTape Assay for TapeStation Systems Quick Guide

For Research Use Only. Not for use in diagnostic procedures.