Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC
    • LC/MS
    • Life Sciences Reagents
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Real-Time PCR (qPCR)
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
Automated Electrophoresis Portal
  • Agilent Knowledge Portal
Automated Electrophoresis Portal
Articles Intensity Loss for 200 nt Ladder Peak in the Agilent TapeStation RNA Assays
  • Articles
  • Files
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • Automated Electrophoresis Portal Home
  • Automated Electrophoresis Portal Articles
  • Automated Electrophoresis Portal Files
  • More
  • Cancel
  • Created :  10 Nov 2022
  • Modified :  7 Feb 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  Automated Electrophoresis Systems
  • Component :  ScreenTapes and Reagents TapeStation Systems
  • Product Name :  2200 TapeStation System 4150 TapeStation System 4200 TapeStation System High Sensitivity RNA ScreenTape Analysis RNA ScreenTape Analysis
  • Task :  Troubleshooting
Related Articles
  • How to Determine the DV200 of FFPE RNA Samples on the Agilent TapeStation

    This Information Applies To: 4200, 4150 and 2200 TapeStation, TapeStation analysis software A02.02 or higher. Issue: When using the Agilent 4200, 4150 and 2200 TapeStation systems, the DV 200 of FFPE RNA samples can be calculated within the TapeStation analysis software. The DV 200 score is a quality score for evaluating quality of RNA derived from formalin-fixed paraffin-embedded (FFPE) samples established by Illumina Inc. in 2016. It represents the percentage of RNA fragments larger than...
  • Discrepancy Between Agilent RIN and RINᵉ Values

    This Information Applies To: 2100 Bioanalyzer, 4150/4200/2200 TapeStation Issue: Discrepancy between Agilent Bioanalyzer RNA integrity number (RIN) and Agilent TapeStation RNA integrity number equivalent (RIN e ) values Resolution: First, please read the following application notes about RIN and RIN e : Performance of the RNA ScreenTape Assays on 2200 TapeStation System Performance of the RNA ScreenTape Assays on 4150 TapeStation System As can be seen in the application notes...
  • Peak Fronting in NGS Library Smears on Agilent TapeStation

    This Information Applies To: 2200/4200/4250 TapeStation systems, TapeStation D1000 ScreenTape assay, TapeStation High Sensitivity D1000 ScreenTape assay, TapeStation analysis software Issue: A sharp peak appears at the lower end of the NGS library smear in Agilent TapeStation D1000 ScreenTape and High Sensitivity D1000 ScreenTape assays (see Figure 1 ). Figure 1 . Region analysis (blue borders) of an NGS library smear. The red arrow indicates the peak fronting. Resolution: Peak...
  • Agilent TapeStation Software Displays a Different ScreenTape Expiration Date

    This Information Applies To: 4200, 4150 and 2200 TapeStation Systems Issue The expiration date displayed in the Agilent TapeStation controller software is different to the one printed on the ScreenTape device. Resolution The expiration date printed on the ScreenTape device applies until the device is removed from its packaging. Once a ScreenTape device is removed from its protective plastic sleeve, it dries out quicker. As the ScreenTape device dries out, bubbles can form in the gel,...
  • How to Export Agilent TapeStation Logs

    This Information Applies To: 4200, 4150 and 2200 TapeStation instruments, TapeStation analysis software, TapeStation controller software Issue: Technical support require the instrument log files to troubleshoot hardware issues with the Agilent 4200, 4150 and 2200 TapeStation systems. It is possible to export the logs via the TapeStation controller or analysis software. Steps to follow Tip : Log files do not include data files, experimental, or sample information. Collect and...

Intensity Loss for 200 nt Ladder Peak in the Agilent TapeStation RNA Assays

Answer

This Information Applies To: 2200/4150/4200 TapeStation systems. RNA ScreenTape and High Sensitivity RNA ScreenTape assays,TapeStation analysis software


Issue:

The 200 nt peak of the Agilent RNA ScreenTape and High Sensitivity RNA ScreenTape Ladder has lost its normal intensity. An example of this phenomenon is shown in Figure 1 with the actual 200 nt peak highlighted with a red circle:

 

Figure 1. RNA Ladder with rudimentary 200 nt peak in red circle

 

If the peak is below a certain intensity, the software will not assign it automatically. Instead, the next ladder peak, which is actually 500 nt in size, will be assigned incorrectly as the 200 nt peak (see Figure 1). You will also see an alert in the Sample Table indicating a missing peak (see Figure 2):

 

Figure 2. Sample Table alert indicating an issue with ladder peak detection


Resolution: 

The 200 nt peak loses intensity, if:

  • Room temperature is too high, near the upper limit of 30 °C for the RNA ScreenTape assays (operating temperature for the assays: 14–30°C). In this case, run the assay again in lower temperature.
  • Heat-denaturation was not effective or not done at all. Heat-denature the ladder again according to instructions in the Quick Guide (72 °C for 3 min, place on ice immediately for 2 min, see also RNA ScreenTape Quick Guide, or High Sensitivity RNA ScreenTape Quick Guide). Ensure that the tubes used fit to your heat block or thermal cycler and that the heat block or thermal cycler is working. 

If the 200 nt peak is not assigned in the software, it has no impact on quantification and RINe calculation as these parameters are calculated independently from the ladder. Quantification is based on the known concentration of the lower marker. The RINe algorithm evaluates the eukaryotic 18S or prokaryotic 16S rRNA signal against the fast region of the electropherogram (region between the 18S or 16S rRNA peak and the 5S region). However, if you have a rudimentary signal, you may assign a 200 nt peak manually. This change is needed, if you want to do DV200 analysis and need to add regions correctly, or want to resolve the alert shown in Figure 2:

  • In the electropherogram or the gel image, right-click the peak or band area and select Add Peak (see Figure 3):

 

Figure 3. Right-click opens the electropherogram/gel context menu.

  • The software will automatically assign it as 200 nt peak and the alert will be gone. However, the Observations column in the sample table will show Sizing Changed (see Figure 4): 

 

Figure 4. Sizing Changed observation

 Tip 
Learn how to effectively operate your Agilent TapeStation RNA ScreenTape and High Sensitivity RNA ScreenTape assays:
Agilent RNA ScreenTape Quick Guide for TapeStation Systems
Agilent High Sensitivity RNA ScreenTape Quick Guide for TapeStation Systems

 

For Research Use Only. Not for use in diagnostic procedures.

 

PR7000-7925

 