Calibration table not accepting retention time adjustments after reprocessing

Hello. I've started having an issue where chemstation doesn't save my retention time adjustments when I reprocess the calibration sequence. I'm including screenshots and i'll walk through this in steps.

1.

Step 1: Run the calibration sequence and open the data file. We'll track glucose for this explanation- notice that the differences in actual vs. expected retention time.

2.

Step 2: Adjust retention times to match observed RTs. I can click around to other samples in the calibration sequence and the times remain adjusted.

3.

Step 3: Reprocess- this screenshot was taken while the sequence was reprocessing. The adjusted times show up in the calibration table as the sequence is reprocessing, however....

4.

4: The times go right back to where they were before I adjusted them.

Does anybody have any insight into this? We run 10 HPLCs with this same method and I've only encountered this problem with this specific instrument.

Thanks in advance!!!

Parents Reply Children
No Data
Was this helpful?