I have ran into another issue when trying to install openlab 2.7. Any suggestion? Thanks
I have ran into another issue when trying to install openlab 2.7. Any suggestion? Thanks
Hello,
New install or upgrade? Typically, when the software install gets stuck in caching, it is an issue with the downloaded media or root certificates. Update the root certificates on your computer or download the media again if it has not been used successfully.
Marty
Upgrading from Chemstation to CDS 2.7
In my experience, Chemstation and CDS do not like to co-exist. You likely need to uninstall Chemstation and remove all the registry entries for it also. It may be easier to start with a fresh computer that has not previously had Chemstation installed on it.
Hello,
If you had Chemstation on the computer, it would have made you uninstall it. When you did this was the verification software uninstalled?
Marty
Hello,
If you had Chemstation on the computer, it would have made you uninstall it. When you did this was the verification software uninstalled?
Marty
Thanks Martin, I finally did a clean install of windows and installed openlab from a fresh computer. Everything is working now. Thank again