Hello! Anyone know why my calibrate curve from Data Analysis are different from excel?
This is my calibrate curve from Vit B3
And this is from excel.
If someone could help me, I will be grateful
Hello! Anyone know why my calibrate curve from Data Analysis are different from excel?
This is my calibrate curve from Vit B3
And this is from excel.
If someone could help me, I will be grateful
Hello,
It is hard to tell without more details but at first glance it looks like you did not include the origin point in the excel calculation. Also, I would need to see the actual values that CDS is using in the calibration curve. If you take the same screenshot but with the calibration table expanded at the bottom we can see the cal injections in the curve with the amt and areas.
Marty
Hello,
It is hard to tell without more details but at first glance it looks like you did not include the origin point in the excel calculation. Also, I would need to see the actual values that CDS is using in the calibration curve. If you take the same screenshot but with the calibration table expanded at the bottom we can see the cal injections in the curve with the amt and areas.
Marty
Here, the calibration table
Hello,
So, I get the same values in excel and CDS when I include the 0,0 point in the excel calculation. I can see in your CDS method and calibration table that you included the origin, so to compare with excel you need to add a 0,0 point to the regression in excel. You also have slightly different area values between the CDS calibration and what you used in excel.
Marty
Ok! Now I see my mistake! I have another question for you, my friend! When I do a calibration curve and I put my amount at level 1, 2, 3... why thoses values are not the same in the injection results amount?
Hello,
Those amounts are recalculated from the calibration curve and the peak area. If you want to see the amount used for the standard curve add the calib amount to the injection results.
Marty
Thank you so much! You are the best!